Check to see if the vertical deflection accuracy is correct.
It may be that one of the horizontal deflection plate leads fell off of the CRT. They should be checked.
Just curious, are those two related? In other words, if the vertical accuracy is or isn't correct does that tell you something about the horizontal deflection leads? I know you are an expert with scope repair. So I am just trying to figure out your troubleshooting thought process to improve my own.
Certainly a good point that the leads could have fallen off! I was going to say measure the voltage range on both leads left and right and see how they compare.
Thank you for your kind words. I do have a lot of experience with analog oscilloscopes.
No, they are not related and disconnecting one horizontal deflection plate should not significantly affect vertical deflection; these are just very easy initial checks which can be made. I did not notice any non-linearity in the video or I would have asked different questions. When the test results are weird, record the amount of weird.
If both the horizontal and vertical deflection are about 1/2 of what they should be, then that indicates either a high cathode acceleration voltage or possibly a low or missing PDA (post deflection acceleration) voltage.
The puzzling thing shown in the video is that the sweep is short covering only half of the horizontal width but the positioning is not centered which would be expected if there was a deflection sensitivity problem. That may indicate a problem with the sweep itself instead of deflection. Update: a disconnected horizontal deflection plate can explain this when the disconnected plate charges to equilibrium.
To try to resolve this without invasive diagnosis, I would check the horizontal and vertical deflection in X-Y mode so the sweep is not involved. I cannot tell from the video if the EAS-201 has a B timebase but if so, I would check its sweep length also.
It would be helpful to see a triangle wave displayed so that linearity errors would be visible as the position controls are adjusted.