Author Topic: Another sick Keithley 2000  (Read 1129 times)
Offline jchw4Topic starter
« on: March 08, 2021, 09:22:23 am »
Some time ago I got mechanically damaged k2000, which otherwise was perfect. It still reads 5.00000 V from my DMMCheck.
(k2000 reports last calibrated in 2002)
But yesterday I noticed that it's 1M range is broken both in 2W and 4W.
For example, 100k in 2W mode:
100k range: 100.7717k
  1M range: Overflow
 10M range: 00.10071M
100M range: 000.1007M
Shorts:
100k range: 000.0003k
 1M  range: -0.000006M
10M  range: 00.00003M
4W:
1.80 Ohm on 1M  : -0.000053 M
2.78 Ohm on 1M  : -0.000072 M
11.8 Ohm on 1M  : -0.000672 M
101.8 Ohm on 1M : -0.002190 M
1.000 k  on 1M  : -0.021508 M
9.967 k on 1M   : -0.214120 M
100 k on 1M     : Overflow
(values on the left measured using other ranges)
All tests pass.
Current source is perfect (measured with external meter).
TP104 seems to be correct.
What could it be? I have a feeeling that it's calibration memory.
Here  https://www.eevblog.com/forum/repair/first-time-on-keithley-2000/msg974309/#msg974309 another user shared calibration dump commands and their actual values.
Here are mine:
000000000393216
000000000196608
000000000262144
+2.74088039E-02
+2.74717795E-01
+2.74241546E-02
+2.74939096E-01
+7.83504339E-05
+2.74746629E-02
+2.75300742E-01
+8.12454578E-05
+2.74820894E-02
+2.75495528E-01
+2.74714668E-02
+1.92738696E-01
+7.83581243E-05
+2.73505044E-02
+2.74136247E-01
+1.37793306E-02
+2.76778882E-01
+9.98209300E-01
+9.52346951E-01
+1.00000000E+03
+9.99984500E+02
+1.00000500E+01
-1.00000700E+01
+1.00001400E+02
+1.00000000E+03
+1.00000000E+04
+1.00000000E+05
+1.00000000E+06
+1.00000000E-02
+1.00000000E-01
+1.00000000E+00
+1.99251556E-06
+1.88449784E-06
-6.65748151E-07
+1.41637118E+00
-1.41637561E+00
+1.41689936E-01
-1.44000000E-01
+2.48186119E-05
+1.72858864E-05
+5.29676357E-06
+3.29774682E-06
+3.18277699E-06
+9.17269493E-02
+9.99929726E-01
-1.93286724E-06
-2.00583545E-06
-1.94541648E-06
-1.97593550E-06
+1.17605155E-06
+1.41526718E-01
+1.42849062E-01
+1.42880911E-01
-6.65767966E-02
+9.17000000E-02
+1.19220998E-06
+1.18602474E-06
+1.43014682E-02
+1.41262068E-03
+1.41139930E-02
Comparing them I can see that one value is significantly  different:
300.000%
-0.537%
-0.261%
-0.712%
-0.403%
-80.024%
-0.480%
-0.216%
-79.150%
-0.664%
-0.360%
-0.586%
-0.288%
-80.008%
-1.071%
-0.791%
-1.431%
-0.824%
-0.179%
-4.247%
0.000%
-0.002%
0.000%
0.001%
0.001%
0.001%
0.000%
0.001%
0.003%
0.000%
0.000%
0.000%
78.899%
130.010%
-4495.288% (-6.65748151E-07   +1.51468621E-08 1111111111111111111111111111111110110101001100101011010111011101 110010100000100001110001001101)
-0.235%
-0.235%
-0.238%
0.000%
61.263%
41.971%
113.748%
186.489%
193.640%
0.090%
0.007%
-24.512%
-17.552%
-23.834%
-22.502%
-2.625%
-0.058%
-0.016%
0.001%
-104.660%
0.000%
104.454%
96.454%
-0.322%
-0.692%
-0.737%
(I am not sure that interpreting these values as float32 is correct, but it looks reasonable to me.)
So my value is -6.65748151E-07 while other users value is +1.51468621E-08 .
Could it be cal EEPROM fault?
What else should I check?
« Last Edit: March 08, 2021, 10:28:33 am by jchw4 »
Online Kleinstein
Re: Another sick Keithley 2000
« Reply #1 on: March 08, 2021, 09:41:54 am »
A negative Ohms readings may be a faulty cal constant. The value seem to be off by a factor -21.
A first point to test would be to measure the test current the meter uses for the tests with a 2nd working meter.
Another possible test would be to measure the "2 wire resistance" of a 1.5 V battery in the 2 M range. This should show how the measured voltage is converted to the result.
The following users thanked this post: jchw4
« Reply #2 on: March 08, 2021, 10:27:52 am »
121GW displays 10uA in 100k and 1M ranges, and 0.7uA in 100M range. It matches k2000 specs.
Measuring Ohms of 1.00000V  source (Ronan X85 calibrator that I posted):
10MOhm: 1.63400 M
1MOhm: Overflow
100k: 99.1294 K
Increasing voltage actually makes ohms read more negative. -1.000098 MOhm matches 0.470060V on the input.
Online Kleinstein
« Reply #3 on: March 08, 2021, 10:34:35 am »
This really looks like a bad calibration constant.  AFAIK the Keithley meters allow to calibrate individual ranges. So one may be able to get at least a somewhat working state back with cal ran for the 2M range.
« Reply #4 on: March 08, 2021, 11:40:27 am »
And I guess that EEPROM needs to be replaced ASAP.
Online Kleinstein
« Reply #5 on: March 08, 2021, 01:53:51 pm »
It is not sure that this is really a damage to eeprom. It could as well be just a soft error, that would be corrected with new written values.
Damaged EEPROMs are relatively rare.
The wrong values could be by an unintended call of a memory write - possibly by the former owner, or a general computer problem like a brown out. In this case even check sums would not help.
Ideally I would expect quite some error correction / checksums, so that corrupted cal data would be rate. The memory chip should have several times the capacity to allow redundant storrage.
At least a good check sum should be there to give a warning.
With computers ideal and real implementation are 2 things.