Hello everyone,
I decided not to go to the Testing Equipment forum with my likely to be trivial question, and searching didn't yield any related result.
I have recently bought a Tektronix 465 (not B) oscilloscope. It was in a rather good condition, written-off military equipment (last calibration sticker is from 2004). Power cord was missing, so I made a new one. Probes were missing too and I bought a pair of Hantek P6100 100MHz ones.
After setting up the capacitance of probes and stuff, I was pleased to see the traces nice and clean. However, there are slight (like 1% of amplitude) overshoots still visible from calibrator output.
Other issue concerns me a bit more. When I move up the vertical position and trace crosses 100% mark, it becomes very fuzzy and somehow gets reflected down a bit (like to a 60% value). After that I can move that fuzzy trace all the way up. If I move it back down, it becomes clear again once the top of the signal crosses 100% mark.
What could it be? CRT failure or something with the electronics? I have already got the service manual, but any advices on where to start looking are much appreciated.