Sorry if this is seen as clutter, two threads for the same (similar?) scope.
I thought if I hijacked the other things would only get confusing.
Anyhoo...
Had this scope for a while, but I haven't used it until I wanted to check a function generator.
After an hour or so the screen went crazy, like the input was being dragged around by low freq signal.
I measured PSU before and after fault condition:
The -15 is a little low at 14.779
More tellingly the +5 is at 4.985v when running, but drops to 4.135Vdc during fault and I was able to measure 1.0768VAC on there too.
I read a few of the threads, and the bridge and filter cap are a big suspect in this fault, with a bridge fault producing weird things on the CRT.
Here's a pic of the screen when faulty (apologies if I cock the size of the pic up):

I have no other scope to check things with, and the option forward seems to remove the bridge and test it.
The 5v bridge is right down against the board (unlike the others, lay on side), very difficult to get to.
I am tempted to cut it out if possible and replace, and whilst there change the cap too.
In that respect, would this bridge and cap be suitable?:
http://uk.rs-online.com/web/p/products/7328257/http://uk.rs-online.com/web/p/products/8116169/Finally, please note I have not worked on scope before, so all warnings are greatly appreciated.
Thanks.