I've been looking at posts on the old Tek 212. My intention was to remove the long dead battery packs to guard against leaking but now it looks like I may have a repair to do. First, I need to determine if the dead batteries or removing the batteries could result in the attached wonky waveform (no input)? Are the batteries performing a filtering function under normal circumstances? Could they be replaced with some electrolytics if I don't intend to use it battery powered? If not the batteries, do you agree that a problem exists in both the horizontal and vertical circuits or before? Traces don't normally fold back on themselves! I did get my hands on the service manual, and I do intend to DEox all of the contacts. I'm looking for a knowledgeable hint from you guys for the best way to start. Perhaps someone has seen this situation before? TIA Rick
 

Offline edavid

  • Super Contributor
  • ***
  • Posts: 3447
  • Country: us
Re: Another visit to the little Tektronix 212
« Reply #1 on: December 28, 2022, 08:41:49 pm »
It looks like the CRT was damaged when the scope was dropped  >:(
 

Offline SuchataviTopic starter

  • Newbie
  • Posts: 2
  • Country: us
Re: Another visit to the little Tektronix 212
« Reply #2 on: December 28, 2022, 09:06:47 pm »
Oh snap that's a shame. Internally it's like new.
 

Offline inse

  • Frequent Contributor
  • **
  • Posts: 824
  • Country: de
Re: Another visit to the little Tektronix 212
« Reply #3 on: December 28, 2022, 09:21:28 pm »
There is no filter cap in the snippet you provided.
Probably the batteries serve as filter caps.
Supply the scope from a lab psu at the battery connector.
« Last Edit: December 29, 2022, 04:04:21 am by inse »
 

Offline Andy Watson

  • Super Contributor
  • ***
  • Posts: 2130
Re: Another visit to the little Tektronix 212
« Reply #4 on: December 28, 2022, 09:29:07 pm »
I cannot see any significant mains filtering apart from the batteries - I would find a suitable DC power supply to check it with.

Also, note that the input, including the battery, is not isolated from the mains.
 

Offline inse

  • Frequent Contributor
  • **
  • Posts: 824
  • Country: de
Re: Another visit to the little Tektronix 212
« Reply #5 on: December 28, 2022, 10:08:18 pm »
On second thought, this might not be the entire power supply of the scope.
The battery is charged via a capacitive dropper which hardly can supply the current required to run the scope.
 

Offline oz2cpu

  • Frequent Contributor
  • **
  • Posts: 882
  • Country: dk
    • webx.dk private hobby and diy stuff
Re: Another visit to the little Tektronix 212
« Reply #6 on: December 28, 2022, 11:57:02 pm »
wow you got the 212, those tiny ultra rare tek scopes are fantastic, i feel you pain mate,
I got a 213 with a blown input dual fet, no suggestions yet about a replacement..
maybe i can chear you up with a little video of mine :
https://youtu.be/l8WaflwgxMs
it is ofcourse click bait, so only click responsible.
Radioamateur OZ2CPU, Senior EE at Prevas
EMC RF SMPS SI PCB LAYOUT and all that stuff.
youtube : oz2cpu teardown
 

Offline wasedadoc

  • Super Contributor
  • ***
  • Posts: 1673
  • Country: gb
Re: Another visit to the little Tektronix 212
« Reply #7 on: December 29, 2022, 12:10:07 am »
It looks like the CRT was damaged when the scope was dropped  >:(
Doesn't look like a damaged CRT to me. My first suspect would be power rail(s).
 

Offline Andy Watson

  • Super Contributor
  • ***
  • Posts: 2130
Re: Another visit to the little Tektronix 212
« Reply #8 on: December 29, 2022, 12:11:30 am »
Here's a little bit more of the power supply circuit. The only significant capacitance I can see is the 100uF on the centre-tap of the transformer. I would definitely find another source of power, or batteries before condeming it.
 

Offline inse

  • Frequent Contributor
  • **
  • Posts: 824
  • Country: de
Re: Another visit to the little Tektronix 212
« Reply #9 on: December 29, 2022, 06:30:37 am »
I am still puzzeled by the power supply.
The two capacitive droppers C210 and C212 in parallel deliver 225mA at 115V/60Hz.
This may be sufficient to charge the batteries, but not to operate a scope with 2.7W of available power?!?
« Last Edit: December 29, 2022, 06:34:12 am by inse »
 

Offline med6753

  • Super Contributor
  • ***
  • Posts: 11329
  • Country: us
  • Tek nut
Re: Another visit to the little Tektronix 212
« Reply #10 on: December 29, 2022, 11:25:08 am »
The batteries need to be installed and functional in order for the PSU to operate properly. Chances are the original NiCd cells look like this:



I decided to replace them with 6V NiMH cells and surprisingly they work fine without any other circuit modifications. Check connector P3 for +69V, +5.6V, and -5.6V. If OK then your issue lies elsewhere. And unfortunately to me at first glance it appears that the CRT might indeed be damaged. Analysis of the waveforms shown in the manual should prove/disprove that once you insure the PSU is OK.


 

And yes, the primary side of the PSU is not isolated from the mains so be cautious of that. Re-assembly of one of these mini scopes can be tricky and frustrating. I've talked with former Tek service techs and they hated working on them.