Self cal doesn't do much, either mixers are completely dead or either something is really strange.
Basically my current situation is just crazy:
- Did "All Cal", it went fine. And actually I've checked what happens if you just remove cal input from main attenuator - the device doesn't care, there is simply no "calibration failed" errors (or I was not able to trigger them any way possible)
- Did preset -> All, and I'm getting straight line over 0 dBm on whole spectrum
- Device reports >+4 dBm (sometimes +300 dBm) over all spectrum
- If I provide some input signal, spectrum around this signal jumps down to ~-50 dBm
- Noise level (can you really even say +4 dBm is a noise level???) jumps between different levels when I switch between different RBW.
- Backup battery is doing fine, minor level of oxidation is present though.
Not having tools to properly test 1st and 2nd mixers (I don't even understand wiring of them) makes me sad
For now I just want to know what works and what doesn't.