Anritsu spectrum analyzer MS2661 tears
« on: March 07, 2015, 05:51:15 pm »
hi , i am in trouble with this instrument , it power on i hear relays clicking but nothing on display , in stand-by mode since
the problem begin its noisy ..whistling opening the psu have child board with red led (V12 label) flashing fast , dont have
the schematic to test voltages on board no silkscreen labels  , if someone expert anritsu can help me , thanks.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28368
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #1 on: March 08, 2015, 05:20:28 am »
CRT display?

How is the CRT HV derived? From a SMPS? Maybe that has failed?

Hold a scope probe close to the SMPS and see if there is any EMI that indicates it might be working.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #2 on: March 08, 2015, 07:11:09 am »
this model have lcd screen NEC NL3224AC35-01 model
 

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #3 on: March 08, 2015, 10:59:14 am »
I HEARD ABOUT BACKLIGHT ..IF ITS BROKEN THE DISPLAY REMAIN TOTALLY DARK OR CAN VIEW EVEN DARK?
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28368
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #4 on: March 08, 2015, 11:02:42 am »
Try shining a bright light directly at the LCD and if you can see a faint image the backlight only has failed.
Some have simple inverters for the backlight HV, others more complex controllers with auto shutdown in fault conditions.
Again you can use a scope probe like a near field probe to see if the backlight oscillator is working.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #5 on: March 08, 2015, 01:03:30 pm »
HV is present , touching with fingers the output of 2 coils i see sparkling , so you thing ccfl lamp is gone? anyway lighting the panel cant view anything
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #6 on: March 08, 2015, 01:45:36 pm »
Somewhere inside will be a set of test points to measure the power supply voltages. Check these first.

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #7 on: March 08, 2015, 02:40:18 pm »
lcd display with his integrated contoller and hv modules get power suppy and signals  form the control panel trough flat cable
FP has test point all 4 values are ok , +14v , -14v , +5v and 12v , , what i found not correct is the Vcc of some ICs on lcd controller
example  TA8696 toshiba controller from datasheet minimal vcc is 4v but i can measure only 1,98v
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #8 on: March 08, 2015, 02:47:44 pm »
Is something on that 4V rail getting hot?

Offline DaxxinTopic starter

  • Regular Contributor
  • *
  • Posts: 213
Re: Anritsu spectrum analyzer MS2661 tears
« Reply #9 on: March 08, 2015, 04:48:55 pm »
None its all fine i found the problem , the bulb ccfl was cracked , alredy ordered the replacement , thanks for the help  :-+
 