Author Topic: HP3325B Fail 139  (Read 314 times)
Offline Greg272Topic starter
HP3325B Fail 139
« on: November 21, 2024, 01:29:15 pm »
Hello, I wanted to welcome all users of HP measuring devices. I also have a request for help in repairing the 3325B generator. The battery was drained after a long downtime, but after replacing the battery, an error occurred during initialization FAIL 139. Does anyone have information about the causes of the error.
Offline Runco990
Re: HP3325B Fail 139
« Reply #1 on: November 21, 2024, 06:19:52 pm »
You lost all calibration values.  Download the service manual and go through all the special function calibrations.
« Reply #2 on: November 21, 2024, 11:13:34 pm »
Thank you for the tips, I don't know if everything is OK because Fail 139 is displayed and it does not respond to the keyboard, is it possible to enter some standard values ​​to restore keyboard support?
Offline Runco990
« Reply #3 on: November 23, 2024, 12:01:14 am »
keyboard should work, double check your work.