Author Topic: Attempt to repair my Extech LCR-200  (Read 1647 times)
Offline icemachineTopic starter
« on: April 13, 2017, 08:12:46 pm »
I try to repair my Extech LCR-200 witch I have demage.
 I think it has happend diuring measurment charged capacitor in cuircuit few days ago.  :(
At the present LCR-200 consume 190mA at 9V DC.
Base on specyfication it should take approximately 35mA @ 9V DC.
At the present when it is supply from 9V it show Er01 on the display and shutdown itself.
I have discover some parts which are demage diuring optical inspection.
I have replace 5 parts 2x SMD resistor,  2x truehole resitor and one transistor. I marked it by red dot.
It not help.
I have discover that decrease supply voltage form 9V DC to 6.9 V DC  could help me to keep multimeter alive diuring investigation.
It simulate weak battery. On this voltage multimeter do not shout down imadiately.
However it stil consume to much power 216mA @ 6.9V DC and do not measure anything. >:(
I suspect that the problem could be ES51920A chip.
Base on that I have measure temperature of it under power suplly 6.9V @216mA DC.
It has 42C deg after few seconds.
I think ES51920A chip has died and I has looking for replacement for it.
I have already read this topic: https://www.eevblog.com/forum/repair/extech-lcr-200-repair/
Mladen thank you for help! :)
Offline mladen82
  • Country: mon
Re: Attempt to repair my Extech LCR-200
« Reply #1 on: April 15, 2017, 07:05:55 pm »
Hi there,
yes its probably ES51920A
send email to Cyrustek and ask for repair for your LCR200
if you lucky like me they will send you 2 pieces  ;)