The supply was shutting down because the output voltages weren't correct. There is a resistive network with all supplies connected which should sum to zero at pin 2 (BAL SENSE) of the controller IC. If it isn't 0V the IC outputs a stop signal on pin 8 which goes to base of Q1254 to shutdown the supply. This was shutting the supply down, and removal of Q1254 on the Cap/Rectifier board allowed the supplies to run so I can fix the output levels (in theory).
The voltages I observed were
Supply Measured
-50V -50V
+50V +46.9V
-15V -14.58V
+15V +13.8V
+5V +4.275V
-50V is the reference for all the other supplies. The other part of that story is "reference ground" which is connected to chassis ground with a 47 ohm resistor (R1591).
In my supply reference ground was 0.45V below chassis ground which was definitely messing up the output voltages.
Shorting R1591 and readjusting the -50V rail to exactly 50V I see the following:
Nominal Actual
+5V +4.75V
-15V -14.21V
+50V +47.70V
+15V +14.32V
Interestingly all of them are nigh on exactly 95% of what they should be! I'm sure that this fact plus the issue of reference ground being at -0.45V without that shorting link should be telling me what's wrong but I just can't see it.
For reference they should be:
Supply Limits
-50 -49.8 to -50.2
-15 -14.85 to -15.14
+5 +4.9 to +5.1
+15 +14.85 to +15.15
+50 +49.5 to +50.5
The raw supply voltages from the Cap/Rectifier board are: -17.6, +17.5, +54.4, -54.4, +7
David