Author Topic: Looking for Mettler Scale ROM backup  (Read 4254 times)
Offline jamarjan
Re: Looking for Mettler Scale ROM backup
« Reply #50 on: November 21, 2024, 09:06:16 am »
Weight arrived.
After motherboard swap, re-soldering the reference resistors and reprograming momeory
my scale works perfectly.
I calibrated it with 1000g standard.
It is very accurate and the measurements are repeatable.
For future reference I am uploading a memory dump and a photo of the nameplate for the PM6000 scale.
I feel unsatisfied with the scale repair attempt, but on the other hand I learned a lot thanks to everyone who wanted to contribute in this thread.
THANKS!
The following users thanked this post: dustooff
Offline gmac34Topic starter
  • Country: no
« Reply #51 on: November 21, 2024, 12:13:42 pm »
Well, to be fair, in the end you have done exactly what the service manual tell you too do in these situations :P
Could you tell what you needed to do to get the scale programmed, I imagine you started with my copy of the EEPROM, did you just replaced the numbers with those from your scale?
Attached you can find a nice document that tells you how to check a balance, if you want to be sure it's performing within specifications.
Enjoy your Mettler