Just a quick sanity check….have you tested the original IC for behaviour ? Is it truly dead/faulty or is the circuit monitoring its output at fault with incorrect on/off thresholds ?
This was the only component that was marked IC and was not a voltage regulator. Output was stuck at 16mV. Fan was always off.
I desoldered it and tried applying external power to the output. At 3V (measured 2.92V) fan turned on, at 2V (measured 2.0224V) fan turned off.
When you removed this IC, the output pad should have risen to +5V via the resistor and capacitor. AISI, if the capacitor doesn't charge, then it must be leaky, or the resistor must be open.
Your idea was right, it's likely a supervisor.
I cheked that it's output goes low when input is more than 4.7V.
When I removed the IC device did not boot until I pu some low voltage on the IC output pin, so yes, it's voltage supervisor.
Thank you! I soldered it back.
Now, back to the fan problem.
This may be of no relevance to this power supply but one of mine uses the calculated wattage to set the fan speed. There is no heatsink temperature sensor for the fan. The lab power supply knows how much voltage and current is being produced and calculates the wattage. This is then used by the microprocessor to determine how much cooling is needed from the fan and the fan speed is set. I was surprised to see such a cooling routine but it certainly appears to work. There is a safety circuit monitoring the heatsink for over temperature in case the fan port becomes blocked. For those doubting this approach, as I did, I tested my power supply response to load and sure enough, the fan has a stepped speed response and each step comes in at a set wattage. At low wattages the fan is off and the next step is cyclic fan on / fan off. Once higher wattages are reached the fan is constantly on and running fast.
As I say, this may have no relevance to the particular power supply that is the subject of this thread, but I thought I would highlight that some power supplies use unconventional fan control systems.
Fraser
This is an interesting idea, but after I recapped the device I set it to load test. With 30V 3A constant output device significantly heats up, but I did not see fan spinning.
If they calculate wattage, it's definitely broken.
Fan only spins in the very beginning when device performs tests. Actually, as I noticed that resetting main IC spins up the fan, it might be just a side affect that FAN PWM was not initialized while booting.
I have a BK 9130 on the round tuit bench which I've recapped in the past but is now reporting wrong readings.
Happy to take a look at any readings you are after, but I agree with fzabkar that the component is most likely a power supply supervisor/reset chip.
Does it spin the fan? Could you check when it starts the fan? I think just one scenario that I could reproduce would be very useful!
The fan is controlled by the 1Q25 (TIP41C) transistor next to to it. Which is controlled by 1IC12 (TL082CN) that you can find on my analog board photo above the transistor along the left board edge.
Here is quick schematics:
Fan (+) goes to unregulated CH1 12V (before the 7812 regulator).
1R39 input goes directly to the pin 3 of the main IC (measured about 0.4 Ohm between them). So I am pretty sure this fan is directly controlled by the main IC.
But I still don't know the soure for the temperature. I don't see any analog inputs of the main IC (so far all I traced were digital lines). It has a few digital output lines to the analog board, but I haven't traced them yet. Maybe it has digital inputs too? I did not get to the analog board yet, so I don't know.
Upd: If you want to monitor fan PWM there are two convenient vias in the top right corner of the digital board to probe: