Author Topic: bradley ct-471c multimeter  (Read 1057 times)
Offline m3vuvTopic starter
« on: October 20, 2018, 10:23:52 pm »
hi all ive one of these im repairing,just a quick question,on the amplifier board there is an axial capacitor 22pf at 750 volt,it seems sort of open circuit ,ie if i feed 100khz at 150mv in one side i get about 20mv out of the other side,the only thing i have is an air variable,can i use that in its place,in the circuit it looks like a coupling cap,regards m3vuv.
Online Gyro
Re: bradley ct-471c multimeter
« Reply #1 on: October 21, 2018, 09:44:49 am »
Rather a dumb question as you don't indicate the input impedance of your measuring instrument or measurement setup. A 22pF capacitor has an impedance of about 72k at 100kHz.
If you're measuring with a basic dmm with meter leads etc, then you'll get loads of un-quantified parasitic load capacitance that will swamp it, that's certainly going to make it seem 'sort of open circuit'.
It presumably has a 750V rating for a reason.
« Last Edit: October 21, 2018, 09:47:19 am by Gyro »
Best Regards, Chris