Author Topic: Problem - Network Analyser HP8713B  (Read 3977 times)
Offline omega12Topic starter
« on: April 17, 2016, 10:02:20 pm »
Hello,
my Network Analyser HP8713B has problems. You can see on the picture that it has problems. I connected a 50Ohm through cable. Need some ideas to repair it. Don´t have extension boards :( but maybe someone can help me. Sorry for my bad english.
Thanks, Regards
Franz
Offline Jay_Diddy_B
Re: Problem - Network Analyser HP8713B
« Reply #1 on: April 18, 2016, 01:09:12 am »
I have an HP8714C which should be similar. Here are a couple of tests to get you started. The idea is to measure the amplitude of the RF source.
First I set the Sweep to CW and the Freq to 10 MHz:
I then set the level to -20dBm:
I can measure the output on a Spectrum analyzer, set for a centre frequency of 10MHz and 5 MHz SPAN:
Or measured on a scope. The scope input has been set to 50 Ohms:
I get 61mV p-p. (It should be 63.6mV)
Try this test to see if the source is working.
Regards,
Jay_Diddy_B
« Last Edit: April 18, 2016, 01:11:28 am by Jay_Diddy_B »
« Reply #2 on: April 18, 2016, 06:08:24 pm »
Hello Jay_Diddy_B,
thanks for your answer. I made the test and it´s look like the source is ok. See the pic please. CW is at -10 dbm by 10 MHz. I don´t know if its normaly but something sweeps slightly at CW from left to the right but i think this is normaly. If i chose sweep 5 MHz than i can see a beautifull sweep on the spectrum analyser. Hm, next step? :)
Thanks, Franz
Offline Jay_Diddy_B
« Reply #3 on: April 18, 2016, 06:28:24 pm »
Franz,
The next step is to go to the Keysight website and download the Service Manual. Here is the link:
http://literature.cdn.keysight.com/litweb/pdf/08713-90005.pdf?id=828971
Section 5 of the manual describes the test software that is built into the analyzer.
Work through this section and see if the analyzer gives any error codes.
Regards,
Jay_Diddy_B
« Reply #4 on: April 21, 2016, 07:57:54 pm »
Hello,
so, i found the problem. The detector diodes are toasted and i wanna also replace the protection diodes. It´s not easy to find original replaces but i hope someone have a good idea :)
I need 4 x detector diodes HP part number 1900-0293. This are Schottky signal diodes SMT.
I also need 4 x protection diodes HP part number 1900-0273. This are Pin Diodes.
Hm, not an easy job.
Regards, Franz
« Reply #5 on: June 12, 2016, 07:01:15 pm »
Hello,
i found the problem. It was U1 on the A4 Board. After replacing U1 the analyzer is working very fine.
Regards,
Franz