I noticed that DER DE-5000 LCR meter could complete calibration procedure with TL-21 adapter. "OPEN" calibration showed "PASS", but "SHORT" calibration showed "-00-" for a long time and then "FAIL".
On the other hand, tweezers adapter TL-22 calibrated with no issues.
LCR meter is relatively new and works well, I did not expect any electronic damage to meter itself. Obvious suspect was TL-21 crocodile clip adapter.
External continuity check with multimeter did not reveal anything out of ordinary. TL-21 has sense and force terminal connected/shorted at the base of the crocodile clip, so continuity check passed. By the way, TL-22 also has sense and force terminals connected near tweezers tip.
I opened the TL-21 case to check internal connections. Solder joints between right angle PCBs were cracked for "-" red and "-" white terminals. Cracked joints created intermittent connections and failed "SHORTED" calibration. Cracking could occur because of structural stress on these solder joints - they are subjected to both bending and pushing forces.
See close-ups of solder joints. TL-21 adapter without the case is inserted in DER DE-5000 for ease of photography.
Action plan for repair is obvious:
- Remove unleaded solder, clean joints, resolder with leaded solder. This should improve resistance to cracking.
- EXTRA - TL-21 leads are too short, almost too short to be usefull at all. 4 wire clips are in order - upgrade will add more cable length and ability to check to broken joints (because all 4 wires cn be open when clips are open).
Things to keep in mind
- TL-21 and TL-22 adapter construction (including all derivations) is prone to solder joint cracking and intermittent connections. Check and resolder if needed.
- TL-21 and TL-22 adapter construction does not allow to easily check for cracked joints, because sense and force terminals are always connected.