It turns out that I didn’t remove Q3, the transistor was badly soldered back at the Bryman factory, the transistor fell off there and they calibrated the multimeter without it, because when I installed the 2sd2657k transistor, there was such an offset in temperature. It was necessary to calibrate the multimeter for temperature again. Trial was able to do it.
How did i do it-
-Short J1 inside multimeter (otherwise you will not enter the adjustment mode)
-While holding the (Hz) button, turn the switch from the (OFF) position to the (Ohm.buzzer) position, hold the (Hz) button until the display turns off, release the ss button
-300.0 will appear on the display, then with the rotary switch we select what needs to be calibrated, I chose (mV/Temperature)
-Button (range) select modes and ranges, chose (--0C)
-Then I connected the thermocouple, lowered the thermocouple into a glass with melted ice (you need to cool the sensor to 0 degrees Celsius)
-Then press the button (select) the calibration values are displayed on the display, wait for the values to calm down, then press the button (hold) the values saved
-Rotary switch switch to (OFF)mode