Hi bob91343,
Before anything, check the +10 Volt reference. Nearly all front panel potentiometers are supplied from this +10 Volt reference. The outputs of these pots are inputs to ADCs.
When the reference is off, so too will be accuracy and attenuator balance - the latter causing vertical trace shift when changing vertical Volt/division and variable vertical gain.
R1292 on A2 is to be set within 9.99 and 10.01 Vdc. Closely observe the voltage at A2 TP201 for any sign of drift. Nudge R1292 ever so slightly and again note any tendency for drift. I’ve now replaced noisy R1292 on three 24x5 scopes. This Bourns 1 kOhm pot has the proper lead spacing and orientation:
https://www.digikey.com/products/en?keywords=3386S-1-102LFIt should go without saying that all power supply voltages and ripple should be checked, per Table 5.1, section 5 page 2 of the manual.
Regards,
RF+ Tech