I posted this in Jan. Still working on the scope.
{{I have a DSO3202A scope which has the following problem, it will not run the 'Self Calibration' sequence. Continues to display the message "Disconnect all Inputs", even when all inputs are disconnected. What's more, looking at the inputs there is no sign of abnormal voltage or resistance. I suspect the problem is with channel 1. That is where the Self Cal. program starts, and it never advances past trying to get channel 1 input to indicate that there is nothing connected. Also, both channels operate normally with signal, and are right on, spec. wise. The only thing I can find is that when Channel 1 is switched from the 100 mV/div position to the 50 mVdiv position the reference is no longer 0, but goes to a -150mV level, and stays at that level when switching lower to the 20 mV/div position. All other positions of the vertical settings are 'spot on', and the scope appears to operate normally, with AC gain accurate on both channels. Normally I would expect the Self Cal. to correct problem at the lower setting on Ch. 1.......except the Self Cal. will not start because the program 'thinks' there is something connected to one of the inputs, presumably Ch 1.
Of course, it appears impossible to get a circuit diagram for this scope. If wrong please let me know. I would appreciate any help I can get in solving this problem, as the scope appears in good shape other than this problem. I have opened the scope and compared measurements of components in the channel 1 and channel 2 attenuators, and can find nothing different between the two. Cannot power up the scope in the disassembled state to check voltages.
Any ideas greatly appreciated!}}
I now have an operational 3202A that does calibration just fine. Of course it is not possible to get schematics for this scope, and does not appear that I can operate it when disassembled. Here is what I have learned. Almost certainly, the Calibration sequence is starting OK only to sense that there is still something attached to an input, probably channel 1. There is not, the impedance is 1 Meg. and no voltage of any kind. When the cal. sequence starts there is a relay activated, a white square appears in position 1 on the display, then a bit later another relay fires, or tries to, and the message 'disconnect all inputs' appears, or remains, on the display, not sure which. The sound of the relays firing does sound a bit different in the operating scope and the bad one, not sure what that means. It is my guess that a relay is not connecting the cal. signals to the channel input. Does that sound like a valid assumption, anyone? Has anyone out there experienced a similar problem. Any help, or suggestions will be appreciated!