« Reply #2 on: March 02, 2016, 05:52:04 pm »
I have two 87 iii meters which is the next version from the 87.
Your meter is functioning normally.
In capacitance mode , no probes attached, my meters reads .33 to .35nF.
This meter is not easy to use for capacitance testing as per the manual and my experience.
Page 19 and 20 of the manual. The meter only reads to 5uF after that you must set the range and count the seconds to charge capacitor and then compare to chart.
https://physics.ucsd.edu/neurophysics/Manuals/Fluke/Fluke83.pdfThe AC random number display is simply noise.
I really like my Fluke 87's but I believe for capacitance the V series may be easier to use.
« Last Edit: March 03, 2016, 01:39:17 am by Johnny10 »
Logged
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, Fluke 181, uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP8568B