Author Topic: B&K Precision 2125 Oscilloscope Vertical Shift  (Read 728 times)
Offline rsaxvcTopic starter
« on: March 14, 2022, 03:58:34 am »
About a year ago I picked up quite a bit of test equipment, including several dual-trace analog oscilloscopes, a few of which have this issue where the vertical channel shifts up and down for a bit. It doesn't happen reliably, but often happens for a couple minutes on and off. Unless I'm monitoring something - then it works fine.
I've got a short video of it here:
It affects both channels similarly, independent of the vertical scale - I think this means that the problem is either after the preamps in the vertical amp or in something that affects both preamps similarly like a power supply. I figured I'd ask if anyone had seem something similar or might have some direction on where I should look next.
Todo/ideas:
  • check XY mode to see if it affects horizontal amp
  • check low/mid voltage power supplies with a scope - nothing interesting with a DMM
Offline bdunham7
Re: B&K Precision 2125 Oscilloscope Vertical Shift
« Reply #1 on: March 14, 2022, 05:08:46 am »
I'd take the cover off and try carefully wiggling all the various inter-board connections, jumpers and grounds to see if something is a bit loose.
Offline Wallace Gasiewicz
« Reply #2 on: March 14, 2022, 04:47:05 pm »
I second the cleaning of connections and cleaning the pots and inputs.
Maybe even a bad solder joint.
« Reply #3 on: March 16, 2022, 07:16:22 am »
If you checked the PSU voltages and they are OK then concentrate on an issue in the vertical amplifier. That's vertical shift, not horizontal shift.