Bicurico:
Hi,
I don't think that swapping the CF cables would cause any harm - you would just map slot 1 to CF2 and vice-versa.
Unless, of course, you put the flat cable the wron side around, which again I don't think is an easy thing to do...
You might have:
- broken HDD
- corrupted file system
- broken module
Or, and this is a long shot: because you connected the CF slots in the wrong order, the C, D, E, ..., letters are now messed up and the BIOS/filesystem tries to load files from the wrong unit?
Good luck!
Regards,
Vitor
Ice-Tea:
Well, figured it out in the end. There was another small flatcable not fully inserted somewhere in the chasis.Hadn't touched it but must have gotten a bit loose during assembly ;)
For my understanding; there's 3 flatcables for the CF card. How do you come to the conclusion swapping them could not cause physical harm?
Bicurico:
Great news! I am glad it was a simple thing.
Please allow me to apologize - I forgot that the card-reader does feature three flat cables and they indeed all have the same connectors. Plus the length of the cables is ALMOST identical - W216 is shorter. I went to my spare parts box and looked at it.
I would say you could manage to swap W217 and W222.
Anyway, I stand by it that you probably would not break anything if you swapped the cables, as I think that there is no dangerous voltage in any of those pins and ground is the same. Some pins are not even connected, as the resistors are missing.
I suspect the third cable is not even used? The PCB is missing an IC that connects to W216.
Regards,
Vitor
Ice-Tea:
I'd give it a try but I don't want to jinx it ;D
Anyways, using the same cables 3x on the same location is just asking for trouble IMOH.
For those stumbling upon this thread somewhere down the line: atached is a picture of the offending flatcable.
Bicurico:
If I am not mistaken, that is the flatcable that connects the RF-Frontend module to the main board.
It is best accessed from the other side, when you remove the RF-Frontend module (I think).
I can imagine all sorts of errors if this cable is not propperly fitted and yes, it could get loose while messing around with RXTX-modules... (they need a lot of unscrewing of the waveguide connectors).
I wonder how long it took to develop the CMU200, considering its complexity and engineering...
Regards,
Vitor