Author Topic: CMU200 Rohde & Schwarz hangs during BaseDiscoverOptionsBegin  (Read 16465 times)
Offline Ice-Tea
    • Freelance Hardware Engineer
Re: CMU200 Rohde & Schwarz hangs during BaseDiscoverOptionsBegin
« Reply #25 on: April 23, 2019, 04:13:16 pm »
I wonder how long it took to develop the CMU200, considering its complexity and engineering...
Think the better question is "how big was the team" ;-)
Offline Bicurico
  • Country: pt
    • VMA's Satellite Blog
« Reply #26 on: April 23, 2019, 04:31:02 pm »
Yes, of course. I meant "how long" in "man years"...
Still, it is impressive when you are messing inside a CMU200: so many modules, each with several boards, custom IC's, FPGA, software, etc.
Then all the PCB layout, circuit design, materials (silver plated modules, etc.).
I can't begin to imagine how much work, research, development went into such a device.
No wonder they were sold new at around 100.000 Euro (?).
Regards,
Vitor
Offline Ice-Tea
« Reply #27 on: April 23, 2019, 06:00:31 pm »
Pretty sure for 100k you got a base unit and not much else...
Offline Bicurico
  • Country: pt
« Reply #28 on: August 21, 2019, 08:31:03 am »
Yesterday one of my CRTU units booted first fine. Then I switched it off but remembered I needed to measure something else, so I turned it on again. Unfortunately the serial number recognition during BaseDiscoveryOptions took much longer than usual and the dreaded error message appeared.
Strange, though, it said "error# 1095 overlay.c system error" instead of the usual "dsp.c" error.
I restarted and then the CRTU booted OK, but on successive boots, even after one night off, it now always produces this "overlay.c" error.
I don't have the time and nerve right now to start swapping HDD and DIGITAL board, though I am pretty sure this will be DIGITAL board related.
Any idea what this could be?
It disturbs me that these fine units are starting to fail due to the DIGITAL board and the small PCB's attached inside, which are basically impossible to repair.
Regards,
Vitor
« Last Edit: August 21, 2019, 08:43:52 am by Bicurico »
Offline Ice-Tea
« Reply #29 on: August 21, 2019, 09:17:18 am »
If you figure out which board crapped out, contact me. I can probably offer a spare.
Offline Bicurico
  • Country: pt
« Reply #30 on: August 22, 2019, 08:38:16 am »
Update:
Swapped DIGITAL board with a known good unit --> same error message! This is NOT related to the DIGITAL board...
Run scandisk and defrag from Service Menu --> No issues found
Swapped HDD with one of another CRTU --> device booted OK, but with fW 3.x instead of FW 5.x (which was active on that HDD -> because of compatibility with Windows 2000 partition)
Went inside Service Menu and installed "Another FW", selecting an alternative FW5.x version (less options installed) --> device boots OK on the original HDD
Conclusion: I think some file/data got corrupted and reinstalling the FW fixed it?
Strange...
Regards,
Vitor
Offline Bicurico
  • Country: pt
« Reply #31 on: August 22, 2019, 09:01:39 am »
Another update:
It is not the HDD or some file corruption.
Activated the original FW5.x with many options and the error comes up again!
So now I come to the conclusion that this error is related to some hardware option that got broken. If the FW does not have support for said option, the unit powers up ok.
Back to FW 5.x with just the GSM option, the device works ok.
Since I swapped the DIGITAL board, it must be some other board/module.
This is a 83 unit. The other broken unit I have is a 82 unit, so I cannot swap all boards, as they are different. I don't want mess with the other working units (CMU200 and CRTU 82).
I don't suspect the RXTX board at this moment.
However, I would be grateful for any idea regarding which board/module might be malfunctioning.
As I really only use this CRTU for spectrum and signal generator, I will leave it as is for now until new ideas come up.
Regards,
Vitor
Offline niconiconi
  • Country: cn
« Reply #32 on: April 28, 2020, 02:38:55 am »
I also encountered a similar problem, and the TL;DL version is: Be aware of the bad IDE ribbon cable!
While replacing the HDD with an SSD, I encountered the hang-on-boot issue on my CMU200. If the hard drive bracket was not screwed in, the boot always succeeded. But as soon as I screwed the bracket in, I had a hang-on-boot problem, sometimes with random system error messages!  :wtf: :wtf: :wtf: It turned out that the SSD enclosure was little bit shorter (or longer? I'm not sure) than the equivalent HDD, so when the hard drive bracket was fully mounted, it put a lot of stress on the IDE ribbon cable and bent it to a sharper angle, causing an intermittent electrical connection, creating the random hang-on-boot issue. If anyone else has similar issues, I think it's a good idea to also check the IDE cable - one way to test it - unscrew the hard drive bracket, flip the HDD, and let the HDD sit on top of the screen with the PCB pointing up (be careful not to accidentally hit or short the HDD while it's spinning), and see if the boot succeeds. If so, the IDE cable is bad. If system failure is mysteriously fixed after removing and reinstalling it back, this can be a reason.
Another issue I encountered was a black-screen-of-death problem when entering Version Manager (even with the original disk), it really drove me crazy - I don't have any of the original installation medium! Scandisk didn't help. But my PC intuition told me it was likely a software compatibility problem, and it could be fixed by using the most ancient way to access the disk. First I entered BIOS and tried setting the HDD access mode from Auto to CHS, and set its speed to PIO 0, the slowest configuration possible. But it just led to an error on boot - CHS is too ancient, so I changed CHS to LBA. And the Version Manager came back to life. So, LBA + PIO 0 is also worth trying.  :-+ The side-effect is an extremely slow disk access, but it can always be changed back after troubleshooting.
« Last Edit: April 28, 2020, 02:46:37 am by niconiconi »
Offline Radiosonde
  • Country: at
« Reply #33 on: December 23, 2020, 05:58:53 pm »
Hello
I now have a CMU200 with a similar problem:
The unit boots up, which it does really slow, if I want to go into "Version Manager" it just shows me a black screen until i turn it off again.
If the unit comes on eventually everything seems to work, altough the cursor sometimes lags, but when I turn on "spectrum analyzer" the CMU shows me a red warning which says:
System Error
Filename: tsk.cpp
Line: 2146
Add. Info
Str: ISR_C:HWIRQ::get_actual_stack_frame_number()
Can anyone give a hint on what this exactly means?
Help would be very much appreciated.
Offline Ice-Tea
« Reply #34 on: December 23, 2020, 07:47:44 pm »
With such erratic behaviour I would firstly:
- check all ribbon cables. A lot of people have been bitten by badly inserted cables, and there are a lot of them
- the "computer board" (which is right behind the display) sometimes suffers from worn out capacitors. Check them. That would fit rather well with bad boot behaviour.
Offline richnormand
« Reply #35 on: December 23, 2020, 10:29:04 pm »
"sometimes suffers from worn out capacitors. Check them."
That would be appropriate indeed.
On mine the small fan was completely seized up. I had to test all the caps but only found one that was marginal. There was a tell-tale temperature sticker on one of the chip that was indicating an overtemperature to its max indication.
Also some of the ribbon connectors were at an angle. So I systematically reseated everything on the whole unit that I could get my hands on.
Repair, Renew, Reuse, Recycle, Rebuild, Reduce, Recover, Repurpose, Restore, Refurbish, Recondition, Renovate
Offline Radiosonde
  • Country: at
« Reply #36 on: December 31, 2020, 10:12:46 pm »
Thanks for the hints so far, and a happy new year all around!
I did so far:
Cleaned all modules and checked all connectors
Replaced BIOS battery etc.
I found that one module on the DIG board causes the CMU to hick up, as described above.
I started with removing all extra modules from the DIG board, now the CMU boots up quick as normal an even passes selftest, as long as it doesnt get to a "spectrum analyzer" function, here it will show me an error "dsp.cpp" "ddc_drv" or similar.
According to the service manual the DIG board hosts the following modules:
-ADC
-DDC
-TXDSP
-AUC (directly above TXDSP with the cooling mat on it, havent marked it in the image)
it is possible to have each module twice as an option, therefore the extra connectors.
The unit boots up normally only if the DDC module isnt fitted, otherwise it will take an eternity..
As said, without the DDC module everything seems fine until I start the spectrum analyzer or the unit itself tries to do so (selftest) which seems logical as the DDC module is there to preprocess the datastream it gets from the ADC.
From the service manual: (courtesy of Rohde & Schwarz)
"The sandwich DDC MODULE1 is directly plugged onto the DIGITAL
BOARD via three multipoint connectors and processes the digital data
stream of the receiver. In a special ASIC chip, the I/Q shaping, the
matching of the data rate and the respective filtering (bandwidth
shaping) of the digital data stream are produced. Then follows a DSP
(RX DSP) with further evaluation of the digital I/Q data for
measurement purposes. Besides, the digital I/Q data are passed on via
the MOTHERBOARD1 to the LINKHANDLER MODULE for evaluation."
I started to reverse engineer the module, and checked all the supply voltages which seem to be correct.
So I guess this isnt repairable at all, at least not without schematics and programmed spare parts.
What would the experts do?
Change the whole DIG assembly or is there a way to get hold of just the DDC board?
Regards
« Last Edit: December 31, 2020, 10:15:14 pm by Radiosonde »
Offline sardonyx
  • Country: ua
    • my blog
« Reply #37 on: August 19, 2023, 11:15:18 am »
Thanks RF_fanatic. I recently bought this tester. It looks like I have the same problem with the TXDSP module. CMU200 hung during BaseDiscoverOptionsBegin for about 5 minutes. I disassembled and cleaned the DIG board. And then by the method of elimination, I realized that the matter was in the indicated board. Without it, the device boots immediately. But now the problem is that the signal generator does not work (it did not work before).
Initially, I thought that the problem was in broken blocks on, or it was in the firmware. But it looks like a hardware problem.
I have not yet found where to buy a replacement board. And I want to try to find out what is wrong with it and maybe try to fix it. Does anyone have any ideas on this where to start or experience with a successful repair?
There are definitely no BGA chips on it. I looked at the board under a microscope - all the elements and connections are in order. At first glance, everything looks working.
« Last Edit: March 15, 2024, 12:03:31 am by sardonyx »
Offline sardonyx
  • Country: ua
    • my blog
« Reply #38 on: March 14, 2024, 11:56:33 pm »
My attempt to repair the DIG-TXDSP board
I bought a CH341a programmer and tried to read the EEPROM with a clip - it was read without problems. From which you can see the following information.
DIG-TXDSP
VER ALTERA=05.01
BD_RAM_K=0256
DS_CKMHZ=100
DS_TYPEID=56303
DS_MASKNR=0K36ADS_HASVSL
Judging by the fact that this information corresponds to the board, I concluded that it is read correctly, there are no reading errors and there is no point in changing the EEPROM. Also, the most likely reason, as it seemed to me, is the failure of the SRAM. My CMU200 worked for more than 15k hours. And it is likely that the number of read/write cycles of SRAM has exceeded a certain limit. After which the crashes begin. Having made this assumption I decided to change all the memory chips on the board. They are relatively inexpensive. But I didn’t want to guess which one had failed and how long the other chips would last. Therefore, I ordered 6 microcircuits from Ali and replaced them immediately.
I must to say that original chip was CY7C1019CV33-10VC and I bought almost the same CY7C1019CV33-12VC. The only difference is a time for rising edge. For the first one it is 10ns, for the second one it's 12ns. I didn't know would it will work or not, but I dicided to risk.
To replace it, I also purchased an Aoyue 1257 hot air nozzle to heat only the ICs contacts. Aluminum foil were also useful to protect the plastic elements from overheating. To solder new SRAM ICs to PCB I used 1.0 mm microwave solder tip.
But these actions had no effect. Any ideas?
Offline Bicurico
  • Country: pt
« Reply #39 on: March 15, 2024, 08:44:57 am »
Congratulation for your efforts, I am reading this with great interest, as I have several of these boards broken.
Unfortunately, I fear that the culprit is the Altera FPGA. Not sure if these are pre-programmed or if their FW is loaded during boot. This is just a guess, though.
The following users thanked this post: sardonyx
Offline sardonyx
  • Country: ua
    • my blog
« Reply #40 on: March 15, 2024, 05:40:29 pm »
I have a few questions. Does anyone know the pinout of the main connectors and two additional unpinned ones on this board? Is it possible to move the TXDSP board on the main board to the left or right? It seemed to me that this was possible. But I haven't tried it yet
How do you know if the FPGA is showing signs of life? (I've never encountered them). The complexity of all debugging manipulations lies in the fact that there are no extension cords to place the DIG board on the table...
Offline RaspII
« Reply #41 on: January 15, 2025, 08:52:55 am »
    Hello all,
    hope this post is still active.
    I also have the boot problem (Error #1095, overlay.c) sporadically.
    Sporadically means if the error is available I have to boot for up to 10 times until error is gone.
    If the error is not avialable I can boot several times without any issue.
    So I checked IDE cable, but no issue there.
    Some more information:
    • I own this device since 2024-12-20
  • the error occured the first time after replacing HDD with SDD. So my idea was to swap SDD with HDD again. First tests with replaced HDD shows the CMU-200 boots without any error (but only few tests done)
Im wondering about the HDD itself. It's an Fujitsu MHW2060AT which should have the size of 60GByte.
But when connecting this HDD to a Windows10 PC the system shows 2GB  of HDD is used,  25,94GByte unused. That means about 33GB are missing (not shown?, hidded?).
The 1:1 copied SSD (64GB) shows 2GB used and 57,63 unused (as expected).
I don't have an idea why there is a different usage (HDD/SDD) but I have the feeling this could be the reason for the overlay.c error.
« Last Edit: January 15, 2025, 08:54:35 am by RaspII »
Offline squadchannel
  • Country: jp
  • deepl translate user
« Reply #42 on: January 15, 2025, 09:32:35 am »
Is it Win2K/9x or WinCE?
if it is FAT16, the maximum partition capacity is 2GB, which seems normal. I think the problem maybe bad sectors, etc.
I missed it "bar" photo, it is FAT16, so 2GB is normal.
let's backup with the DD command and then think about it. First, the backup.
if there is a firmware updater for CMU200, it may be possible to extract and restore system from updater file.
Well, I calmed down and looked over this topic.
It is very likely a hardware failure.
If it happens from time to time, it is likely that the problem is not in the HDD, but in the memory, etc.
power supply is also a strong possibility.
« Last Edit: January 15, 2025, 10:21:04 am by squadchannel »
Offline RaspII
« Reply #43 on: January 15, 2025, 10:35:06 am »
I did the backup with the "HDD Raw Copy Tool". What exactly do you mean with the "DD Command" ? Edit: do you mean the Linux DD command? (I did all my work on a Windows10 PC so far)
As far as I know I don't own a firmware updater for CMU200.
In regard to HW issue: yes, the HW is really a candidate.
Nevertheless I'm wondering that I never got this bug with the HDD, originally shipped with the CMU-200 (Up to now I have about 30 boot cycles without any error. Maybe the SDD is more sensible in regard to stable power supply...
« Last Edit: January 15, 2025, 10:57:22 am by RaspII »
Offline RaspII
« Reply #44 on: Yesterday at 04:13:18 pm »
I did some more tests.
Added a new SDD (SanDisk X110 mSATA 32GB ,same mSATA to PATA Converter).
With this SSD every boot ended with Error #1095, overlay.c  (could not achive one ok boot cycle).
After going back to the orignal HDD (see above) no more error.
So I'm now pretty sure, that the SSD itself or mSATA to PATA converter is the root cause for Error #1095, overlay.c.
Does anybody knows what this "overlay.c" is made for?
73