Author Topic: Tektronix 2235 - power supply issues  (Read 4798 times)
Offline akimpowerscr
Re: Tektronix 2235 - power supply issues
« Reply #25 on: October 22, 2020, 09:39:01 pm »
Well done, you have discovered a new fault: cr907 is dead short.
This diode is essential for the operation of the pre-regulator, it will have to be replaced later but no need for it to do the tests with the 43V
No need either for the SCR Q935 (crowbar) or the Q9070.
The consumption of 700mA seems normal to me, the oscilloscope should work.
Check if all the settings on the front panel are correct to have a trace
Offline AaronBTopic starter
« Reply #26 on: October 22, 2020, 11:05:04 pm »
The resistors you mention all look good. For the power rails I have:
TP961   -8.6V
W954    +100V
W956    +30V
W960    +8.6V
W968    +5.1V
I don't have any way of checking the 2KV supply. I started on the trouble shooting flow chart with the initial settings described, and got the above voltage readings, which look correct. Still no base line trace.  But I'm a bit stuck going any further down the trouble shooting flow chart as I don't have another scope to check waveforms. Unless I can do it with one of the digital, ultra-cheap Chinesium, ones off eBay.
« Last Edit: October 22, 2020, 11:25:37 pm by AaronB »
Offline bdunham7
« Reply #27 on: October 22, 2020, 11:39:32 pm »
OK, if you turn the intensity all the way up and press beam find, do you get nothing?  How about in a darkened room?
Offline George Edmonds
« Reply #28 on: October 23, 2020, 12:14:27 am »
Hi
This scope has some 14KV as a Post Deflection Acceleration voltage, if you listen carefully at startup and shut down you should hear a rustling sound as the 14KV comes up and down.
George G6HIG Dover UK
« Reply #29 on: October 23, 2020, 12:17:35 am »
I have played around with various front panel settings and Beam Find, and twice it displayed a brief, fuzzy smudge. But I cannot repeat it. Nothing obvious in the dark either, except a brief flash of the A Trigger LED.
Son of a Diddly. Just as I typed the above, it has sprung into life  ;D ;D ;D
It's a little unstable, with some ripple on CH2 but I'm of to bed happy  :-+
Offline jdragoset
« Reply #30 on: October 23, 2020, 03:36:23 am »
Excellent trouble shooting patience.
Recently I had a 2215 preregulator (A-19) with a very similar problem.
After verifying the rest of the scope was ok (with external 43 volts DC input) I eventually traced the problem to a bad solder joint.
AC line-powered the board externally into an incandescent 75 watt load while flexing the board and eventually found the problem
 :-- I miss the days of no hf, just the line frequency.
Just restored (E-cap change) a Telequipment (Tektronix UK) D 54 with all DC voltages derived from an (impossible to replace) line frequency transformer with multiple input winding from 100 V to 120 V, 5 volt increments or 200 V to 240 V, 10 volt increments, 48-450 Hz.
Two vertical channels and no  B sweep, etc.
One sided PC's, no IC's and all transistors in sockets.
Only 10 MHz bandwidth but good for any audio work.
Only draw back is variable/cal knobs don't have indicators or switch detents and change when changing V/div or time/div >:D
« Reply #31 on: October 23, 2020, 06:29:16 pm »
Although I know have a working scope, in that I can see a couple of beam traces, there are still issues that need fixing. Unfortunately they are intermittent and difficult to diagnose. At anything more than moderate intensity the trace becomes quite unstable, some kind of shearing of the beam pattern, and inability to focus. Also, sometimes there is a tinkling/ringing sound from the HT side of the board at high intensity. I went through the first steps in the adjustment procedure to try and set the grid bias, but it changes each time I turn it on. A couple of times I got a square wave trace from the probe adjust test. And other problems, nothing I adjust seems to stay consistent.
I think I will stick with the external 43V supply - not sure I have the stamina to debug something I can already bypass. I want to isolate the pre-regulator entirely so that I can put the PSU inside the case and use the normal power switch, also to remove any interaction that may be causing problems from the old, unused components. I'd like to do this without cutting tracks, but that may not be possible.
The problem that @jdragoset had with a dry joint may be something similar to my intermittent problems, that circuitry is very sensitive. Last time the scope was misbehaving I found a dead spider under the CRT, removing it fixed the problem.
« Reply #32 on: October 23, 2020, 06:58:19 pm »
Before making any changes, try to repair your oscilloscope pre-regulator.
When a device has been idle for a long time, it is normal to experience bad contacts issues.
On the other hand, cold solder/dry joint problems are quite rare with tin lead solder.
For bad contacts, start by moving all switches  on the front panel a dozen times.
Mark the position of the trimpot sliders, move them around their position then return the sliders to their initial position.
Remove and reinsert all connectors, including the tube socket. (be careful, discharge the 1200V before)
Find out where the high brightness noise is coming from.
Then we will try to fix the pre-regulator
Offline bdunham7
« Reply #33 on: October 23, 2020, 07:35:38 pm »
If you are triple-sure those resistors I mentioned are all good, then you will need to acquire the means to safely measure high voltage before you can do much else diagnosis wise.  The only other way is to examine all of the solder joints and reflow them, measure all of the components (especially the resistors) in the HV section and so on--perhaps you'll get lucky.  The noise sounds like something is breaking down or arcing, but it could be anywhere.  Inspecting under power (carefully) in a dark room may help.
« Reply #34 on: October 23, 2020, 11:17:08 pm »
I appreciate your thoughts but I have to take a more pragmatic approach. The next thing I would need to do to fix the pre-regulator is to source a single diode to get to what may be a fix, or just another round of debugging to find another failed component. Rinse and repeat. Buying individual components piecemeal is slow, expensive and tiresome. I would probably just give up if another single component was the next blocker, without an end game in sight. The external PSU is an easier way forward, and keeps me interested in getting this fix completed.
The intermittent failures I am experiencing can often be induced by just lifting up the chassis frame. To me, that indicates some physical PCB or connection issue is at least part of this.
I had been fiddling with the switches and pots and I think it was that which contributed to it suddenly springing into life. Also, the noise has now gone - don't you love intermittent faults  :(. The most obvious strangeness now is with the trigger LED randomly on/off/flashing depending on the intensity setting.
Thanks for all your help to get me this far. :-+
« Reply #35 on: October 24, 2020, 03:58:00 pm »
To repair these oscilloscopes, you should ideally have:
- a good quality isolation transformer to be able to work and measure the circuits of the pre-regulator in complete safety
- a 0-50V 0- 2A laboratory power supply
- another oscilloscope
- a multimeter
- spare parts
- service manual
- a calibrated function generator
Since you don't have most of these, diagnosis and repair is much more difficult.
That is why we are looking for possible solutions for you.
Regarding the defective diode, you could try to replace it with a UF4007 (UF4007, not 1N4007 !!!) or BYV27 / 600.
For the pre-regulator, I use to supply C925 in 15V with an external power supply and I check the presence of pulses on the gate of Q9070 with an oscilloscope.
It is not possible for you to do this.
So it will take:
- ex officio replace Q9070 and check Q908, CR908, check the PCB tracks around these components, and we will hope U930 is not faulty.
After this, we will test the pre-regulator with an incandescent bulb 100W in serie with mains.
If it try to start, you can feed it directly with mains without the 100W bulb in serie.
NB: before testing the pre-regulator, Q935 MUST BE IN PLACE
« Reply #36 on: October 26, 2020, 11:10:56 am »
- a multimeter
- spare parts
- service manual
Sure, in an ideal world I would have all that stuff.
What I do have:
- a power supply that can do up to 48V at 1.6A
- a kit built digital scope that works OK up to about 50Khz
- regular and clamp multimeters, but only up to CAT II 1000V
- a small stock of resistors and capacitors, and other parts culled from dead equipment
- the service manual
- a couple of signal generators, one audio band, one RF
So it will take:
Spare parts are the main problem. I see that UF4007 is cheap enough on eBay, but I have been burned a few times recently from sellers claiming UK stock with 3 day delivery, when in fact it comes from China in 6 weeks, isn't the correct part, is fake, or doesn't come at all. I can't reliably test Q908/CR908 while on the board, I would have to remove them for testing. Not sure what they are until I pull them out and try to find a cross reference, and a reliable supplier if broken. While probing around that area I find that R909 is open circuit - but I don't see R909 on the schematic or parts list that I have???? So, I'm back to looking at the time consuming, and possibly fruitless, task of debugging and fixing this damn pre-regulator. Or, just plugging in the 43V PSU.
There is some good news: With some more time spent twiddling pots, flipping switches, and knob turning I have found that the intermittent problems I had before have now gone. With the 43v supply, it now works correctly. Or at least as well as it did before the fuse blew. It would be nice if Santa bought me some better test gear for Xmas and I could continue this adventure, but I'll leave things working as they are for now.
« Last Edit: October 26, 2020, 11:22:39 am by AaronB »
« Reply #37 on: October 26, 2020, 12:15:32 pm »
I did not find R909 on the diagram either, but it is probably a 10 or 15 ohm resistor in series with the Q9070 gate to avoid oscillations (to be checked :).
If this is the case, there are probably other defective components in the pre-regulator control circuit and in particular:
Q908, CR908 and U930.
U930 (TL594) will have to be replaced automatically.
Be careful, you must place a good quality socket if you replace U930 !!!!
But you may also use your oscilloscope with an external 43V power supply if you wish.  :-+
« Reply #38 on: October 26, 2020, 02:51:55 pm »
I know there are a number of threads on the problems with this series of Tektronix power supplies, and it looks like there is a bunch of well known suspect components for the possible issues. Can we put together a list of cross referenced components that would cover the most common eventualities? I don't mind ordering a full kit of parts, at a reasonable price, in one shot to save on delivery and processing fees of individual items None of the proposed items in this thread are excessively expensive. That would solve the frustration of waiting for a single new component to be delivered at the high cost of a single item. No doubt you will end up with unused parts, but they can be added to your stash.
The goal would be to work methodically through the test/replace procedure without the frustration of waiting for components. I encountered this scenario when fixing my plasma TV - I found a kit of parts on eBay that contained all the usual failure suspects when that model of TV went bad, and yes, it was always the power supply. The TV broke at the beginning of the Covid lockdown in the UK - quite a lot of pressure from the family to fix it :box: ASAP. The single kit of parts saved me being lynched.
« Reply #39 on: November 01, 2020, 12:28:31 am »
The deed is done. I have tried to post a picture of the mods I did but there is a limit on what I can post (apparently) . Maybe you can see something in this image.
« Last Edit: November 01, 2020, 05:23:11 pm by AaronB »
Offline TERRA Operative
  • Country: jp
  • Voider of warranties
    • Near Far Media Youtube
« Reply #40 on: November 01, 2020, 06:14:23 am »
You posted an image smaller than a thumbnail. :D
Offline ebastler
« Reply #41 on: November 01, 2020, 06:34:26 am »
AaronB originally tried to post a 12 MPixel image, was told in the Moderation Reports thread (where he complained that this didn't work) that such big pictures are usually a bad idea, and now feels entitled to be peeved.
« Reply #42 on: November 01, 2020, 02:58:27 pm »
I did nothing intentionally to mess up this thread or annoy others, I got caught out by - let's be honest - some pretty antiquated software. Can I delete and then resize the previous image I had posted, then reattach a decent sized image of the final build. I don't know. There is not much guidance or obvious warnings for management of attachments - all I saw was a 5MB limit, I tried to upload 3.5Mb and got a security warning - huh?. All the other forum systems I post to are able to handle a phone camera picture, either as a raw image, resizing it on upload, or allowing the poster to choose an image resize. Now I know the limits of this system, I won't make the same mistake.
« Reply #43 on: November 01, 2020, 03:09:26 pm »
You just have to adapt to the forum ...
Anyone can make mistakes at the beginning, you should not make a drama ...
Forget all that and keep posting.  :popcorn:
Offline ebastler
« Reply #44 on: November 01, 2020, 04:39:36 pm »
Attaching that 46*25 pixel image in response to the size limits you encountered came across as a bit juvenile...
You can, of course, change the picture later: Click the "modify" button to the right of your post's title; below the edit window which opens click "attachments and other options"; uncheck the old attachment and select a new picture to attach.
« Reply #45 on: November 01, 2020, 04:57:04 pm »
Childish or not, that's your decision, but it was all I could do. I had already posted one large image, someone else posted 3 other images, which took the total thread size right up to the 5MB limit. What I posted then, was all that was allowed.
« Reply #46 on: November 01, 2020, 05:26:59 pm »
So now it should all be fixed. Two images of a size that fits the download limits of this forum.
Offline ebastler
« Reply #47 on: November 01, 2020, 06:31:57 pm »
No, you misunderstood. The 5MB total attachment size limit applies per post, not for the entire thread.
Edit: I just tried to attach two images here of 1.4 MB each to confirm -- works just fine. Removed the pictures now to conserve server space.
Yes, that means that a single 3.5 MB image attached to your earlier post should have been just fine, assuming that it was the only attachment to that post, and that it indeed had a size below the 4 MB limit for each individual attachment. Not sure what went wrong with your prior attempt.
« Last Edit: November 01, 2020, 06:35:03 pm by ebastler »
« Reply #48 on: November 01, 2020, 07:34:18 pm »
The wording could be construed as a bit ambiguous, and I see now that the limit message has changed to be total size 5000KB, maximum individual size 4000KB. Not sure why my other image was rejected but hey, it's all good now.  :-+ You could do with a :beer: icon
I know it can be difficult to interact with a forum when you are a noob and trying to find your feet. I did search around the internet for a couple of days researching my problem before choosing this site. And I did then search in the Test Gear and the Repair forums before posting here. I'm not sure if I have enough knowledge to contribute with answers to other people's problems, but sometimes it can be enough to further a discussion by just joining in and asking a question.
I very much appreciate the help you have all given, and now have a working 'scope, rather than some more "busted electrical tat" clogging up the place (she said).