Hello EEVBLOG, knew I was going to make an account here one day for repair help, but I did not think the road would start from my own fault.
Short story. Went to measure a voltage during an Arduino project, and while in 4-wire, I grabbed the wrong leads and my 34401A's autoranging started freaking out since I had it in some configuration with my power supply (sourcing 12V). I have been trying to assess the range of the damage and I seem to have gotten away with minimal damage.
I have been working on my repair and circuit analysis skills, but I have yet to tackle something like this.
--------------------------FINDINGS--------------------------
- Self-test passes with no errors.
- OPEN 4-wire resistance measurements now read 0 ohms, or floating sometimes when nothing is connected to the input.
- 1 MOhm and above float, while 100 kOhm and below short out.
However, 2-wire does work properly as it reads open when nothing is connected. The meter is still reading correctly (though the last digits even in 4-wire seem to be a bit more unstable, this could be attributed to the higher temp and humidity in my room right now as well as the meter not being warmed up fully). This is causing the unit to not auto range properly until a load is connected. I did check the source currents for each range using my 87V and it all checks out. I have some L&N standard resistors which agree very close to their last known values. The open and 4-wire short values are open = 3 mOhm, 4-wire short = 6 mOhm. 1 to 3 mOhm sounds about right with the short have been on and stabilized previosuly.
I believe something is primarily wrong with either the circuit protection or sense circuitry. I have thus far checked:
- CR202 - Acts as it should
- Q211 - I do get a 0.7V drop from Gate to drain/source as the drain/source are shorted in the schematic, but nothing with leads reversed. Approx. 600 kOhm reading from G to DS.
- Q201 - Essentially the same as Q211, with gate to drain, and gate to source reading 0.8V, and approximately 500 kOhm between G/D and G/S.
- Q202 - 0.7V drop from source(high) to gate(low), and drain to gate, but not from gate(high) to drain(low) or source. Approx. 450 kOhm from both drain, and source to gate.
- CR201 - When looking top down, so the single pin is at bottom, the right pin has a 0.7V drop, but the left pin is open in both directions. (Possible culprit?). However I can find no datasheet for it.
I am not really sure where to look from here as I have not found any posts with a similar issue as to mine, but it is always possible I was not using the right key-words.
Any help would be appreciated in narrowing this down. I have no experience with SMD components in my soldering ability yet, and nor do I have tools on me apart from an standard tip on a FX888D, so taking anything out to test individually is not ideal, though I understand I may have to.
I do work in a calibration lab, so I can bring this in this week to have the full verification run on it to assess if anything is out of tolerance, but all seems to be working for the most part right now outside the one issue.
Thank you,