Author Topic: Datron 1041 vintage multimeter  (Read 12215 times)
Offline ZeitkindTopic starter
« on: February 02, 2017, 03:45:35 am »
Hi!
On my hobby bench today: A Datron 1041 4-1/2 digits multimeter, some chips date prob. around 1978.
Unit powers up, but display kinda "hangs" at ".9990".
I can select V or Ohms, doesn't matter, display shows ".9990". No click of any relays, but the "+"-symbol changes when I switch from DC to AC. Not dead, but half dead.
I opened the meter, totally unserviced yet, everything looks original. I have the schematics of the big brother model 1051 - which is the very similar 5-1/2 digit one - and the 2 models seem to share some boards. I measured basic stuff like supply voltage (both 5V and 15V seem to be fine) and checked some electrolytic caps (Philips and others, capacity and ESR are all in spec). The EMI filter went up in nasty smoke and my junior got scared :scared: - glad I had a cheap chinese spare. :phew: It took the EMI filter around 10min to blow up, so I think the meter hasn't been powered up for a very long time.
I wonder if anyone out there ever had this problem with a 1041 or 1051?
- tia Hermann
Don't turn it on - it will explode!
Offline Gyro
Re: Datron 1041 vintage multimeter
« Reply #1 on: February 02, 2017, 11:09:57 am »
Hi Zeitkind, welcome to the forum,
Yes, I can hopefully help you. I have a 1041M (mil version) a 1045 (DC only) and a 1030A (RMS) all of which use the same A/D board.
https://www.eevblog.com/forum/chat/show-your-multimeter!/msg793266/#msg793266
Your 1041 is definitely worth persevering with, it has very decent accuracy for a 4 1/2 digit meter (0.01% +1dig DC), a 10mV range with 1uV resolution and the same true RMS converter as the later 106x and 107x ranges (assuming yours has the AC option fitted). The best thing is that it is completely discrete (no micro, s/w or Cal RAM) so you can maintain it almost indefinitely. Nice Panaplex displays too. You can easily change it to display 10 readings/sec too. So don't bin it!
Discovering and replacing the Mains filter is a good first step :-DD, as is checking the three supply rails - the tantalum beads on the outputs of the 15V regulators are also prone to failure. The axial electrolytics seem to last indefinitely so no point in changing them, as you have found. Good that it is displaying something too (+180V, drivers and displays functional).
The 104x and 105x do share most boards (apart from the A/D and display board, there are differences in the quality of ranging resistors etc on the other boards (higher quality on the 105x) but they are interchangeable. Does having the 1051 schematics also mean that you have a 1051?
All the logic ICs are socketed, which will help your debug. The clock oscillator is a simple transistor astable (no pll complication like the 1051), but I suspect that the problem is in the general triggering logic. What colour IC sockets does it have? The translucent poyethene looking ones are less reliable than the black ones.
As the schematics are only available in paper form I will need to get the camera out. In the meantime it would helpful you can post some photos to judge the general condition see if there is anything obvious. Also post which options are fitted.
One other thing, check that all buttons (hold, print inhibit etc) in the centre contol group are not pressed in!
P.S. Does the front panel say 'datron', 'datron Ltd' or 'datron instruments Limited'? This helps to judge the age.
« Last Edit: February 02, 2017, 11:23:06 am by Gyro »
Best Regards, Chris
« Reply #2 on: February 02, 2017, 03:12:43 pm »
Hi, thx for the answer!
Serial number on the back is 3568.. no idea how many they built, "Datron Electronics Limited"
I did a visual inspection, took out the boards, re-seated most of the TLL stuff. Some look .. gunky, grubby. Just as they have a silicone oil like something on them, prob. gonna clean the boards though they look brand new otherwise. All TI 74L90N have black colored pins.. it's a kind of black paint, weird.
When I disconnect the -15V and/or +15V cables to the other boards, the display shows various numbers, so isn't stuck any more. After dis- and reconnecting that rail, I saw smoke coming of a blue tantalum bead cap (C53 on the AC option board). I desoldered it, it has 0F and 24ohms. Other blue tantalums show in-circuit resistance from <1k up to 5M, prob. I should desolder all below some Meg Ohms and check them. I have no spare ones, so I have to wait for delivery now :(
The schematics of the 1051 can be downloaded at http://www.ko4bb.com/ soon, Peter from g0rsq.co.uk was so kind to scan and upload them. I don't own a 1051, but I prob. get one in more or less working condition next week. I took some pictures today outside:
Offline Gyro
« Reply #3 on: February 02, 2017, 04:09:00 pm »
That looks in pretty good condition.  :-+
From the company name and look of the boards, you have a reasonably late one which is good. It looks to be in good original condition, original Wima electrolytics, no obvious signs of anything having been changed. It's good that it has the black IC sockets too, they're pretty reliable.
Be careful of board cleaning and flux generally. There are guarded low leakage areas where you could cause drift. (especially under the black draught shield on the DC board, that's where the factory TC selected and compensated LM312 lives).
Interesting that you now have some activity, it clearly has a few leaky tantalum beads but I wouldn't have expected that to stop the digital section (unless it upset the analogue side of the discrete dual slope A/D converter). Keep a close eye on the -15 and +15V rails. Also note carefully where each connection to the front panel terminal board goes if you remove the AC and Ohms options.
As a start, I've photographed the schematics for the main board which include power supply, A/D and displays. That's the bit that's different from the 1051. It may help get you started.
A warning that the 1051 is rather more temperamental than the 1041, it uses a triple slope A/D conversion rather than dual and its autozero loop is tricky (trying to fix a 1055 board at the moment).
Damn, that means I'm sort of duty bound to start scanning the whole 1041 manual.   It's fairly thick and the schematics are A3 .:palm:
I hope the attached shematics help you start to get your bearings anyway.
Best Regards, Chris
« Reply #4 on: February 02, 2017, 07:09:29 pm »
Thx!!
Without the AC board, the meter shows nothing (blank) on DC, AC or Ohm. I have to wait for the spare parts, and the blue tantals are spread over all boards, most of them show suspicious low resistance - in circuit, but well..
Gonna take a look at the schematics and try to figure out what I should read at the various test point. Some have +4.99V or -4.99V, but on some I see just floating something with my DMM and oscilloscope.
Offline Gyro
« Reply #5 on: February 02, 2017, 07:41:07 pm »
Ah, actually that is progress.
Blank display (apart from decimal point and mV/V/ohm etc) indicates over-range on these datron meters. The Ohms board intercepts the input Lo connection (the ohms board has a relay on-board).
As the DC volts input is high impedance (>10G) it will float to over-range if nothing is connected (input Lo is open as above). You should find a wire loop on the back of the terminal board that is snipped open (it was snipped at the factory when the Ohms option was added). If you temporarily re-make the link and short the input Hi and Lo terminals you should hopefully get a zero reading.
If you can get to this stage then you can check the operation of DC volts ranges, autoranging functionality etc.
P.S. I know a simple mod to make it display a 'leading 1' rather than blank display for overrange - more like the behavior of most DMMs.
« Last Edit: February 02, 2017, 07:44:58 pm by Gyro »
Best Regards, Chris
« Reply #6 on: February 02, 2017, 09:45:23 pm »
Not sure what you call "Ohms board".  I have a "main" AD-board with the mains transformer, the DC Isolator board next to it and the removed AC board with the Datron RMS-Chip (and the blown tantalum cap) that was below the DC Isolator board. If Ohms need an addon board, I don't have one.  :(
I can see a loop behind the Lo(black) input plug, but it is still connected and not cut - because the Ohms board is missing.. *damn*
Connecting a DC source - no display, so there likely is a fault on the DC input board or the main board. But I will replace all those blue tantal caps before I dig further.
And - if I get this unit working, I can measure Volts DC and AC, but no Ohms.  :'(
Offline Gyro
« Reply #7 on: February 02, 2017, 09:57:39 pm »
Sadly, yes your unit does not include the Ohms option. It would have sat next to the AC option (photo of the underside of my 1041M attached, note that the extra cross bracing is for the M version only).
I would concentrate on getting the DC functionality working before plugging in the AC option again. Curious that you had a display before removing it but not now. Is the display completely blank or do you have mV/V indicator and decimalpoint - if not then check the 5V logic supply again.
Also try manual ranging (auto button out) [EDIT: Rem (Remote) button out too!]. Does the decimal point move, mV change to V etc.
« Last Edit: February 02, 2017, 10:15:22 pm by Gyro »
Best Regards, Chris
Offline Gyro
« Reply #8 on: February 03, 2017, 06:49:57 pm »
I've scanned and uploaded my 1041 Service Manual to http://www.ko4bb.com/ today. Hopefully it won't take long to go live.  :)
Best Regards, Chris
« Reply #9 on: February 06, 2017, 10:05:24 pm »
Thx again! Will take a while to go online, but well...
I measured some more of those blue tantalum caps and - well, many of them should be replaced, they have some k ohms on DC or even below. I think about replacing them with some good electrolytics, but not sure..  :-\ .. prob. with a small ceramic disc in parallel?
I have a bunch of smaller tantal caps, but none with 36V .. hmm..
Offline lowimpedance
  • Watts in an ohm?
« Reply #10 on: February 06, 2017, 10:43:40 pm »
Looks like a rather interesting 'old timer', Cant seem to find any pics on the web  :-// hmmm.
Thanks for the internal pics chaps. Nicely built, always nice to see inside these oldies and that they are not forgotten in particular.
« Last Edit: February 06, 2017, 10:47:39 pm by lowimpedance »
  • Watts in an ohm?
« Reply #11 on: February 06, 2017, 10:54:25 pm »
@gyro
 Just checked the linked thread, nice vintage pile there  :-+.
Just had a quick look on that auction site but nothing came up. I assume they must be rather rare beasties then ?. As opposed to the common as mud 106x 108x types  :D.
Offline Gyro
« Reply #12 on: February 06, 2017, 10:57:38 pm »
You're welcome. If you wish, PM me an email address and I'll send it direct (it's only about 3.25MB), your choice though.
I noticed earlier that you mentioned testing the Tantalum capacitors "in circuit".  You won't get a proper leakage reading that way - there is too much other circuitry across the supply rail. Unsoldering them to test will probably kill them anyway so not really an alternative. I would replace the one that has smoked (obviously) but would be tempted to leave the others unless they run warm. The only Tantalums I have seen fail (over several meters) are the ones nearest to the voltage regulators, however your failure on the AC board is an exception.
Yes I would use higher voltage rated caps for any replacements, I think 25V is the next step up from 16V, there is plenty of space anyway. Modern low esr electrolytics should be fine too.
@lowimpedance: Yes, they're nicely built and still very useable meters (well I use mine day to day anyway), 10G input impedance and 0.01% on 4 1/2 digits is pretty much what you read on the display (+/- 1 digit), with a 10mV range they can be more accurate at 10's uV levels than longer scale meters. I'll do a basic teardown of mine (1041, 1045 (DCV), 1030A (RMS AC) if there's interest.
Best Regards, Chris
« Reply #13 on: February 06, 2017, 11:54:39 pm »
Shouldn't be a problem, but I can wait, no need to hurry.
That's quite easy: If I see >1M?, it should be fine. If I see anything below, it might be ok or might be bad - depending on the circuit. So I desoldered one pin of only those that showed a significant low resistance and tested again.
The one that failed is directly connected to the 15V rail and is rated.. 16V. So well, if the rest is as tight as this one, I prob. replace them all. I remember voltage*3 for tantal, so.. 35V? Duh..
On the power PCB I don't see a problem, but I'm yet to understand the AC and DC input circuits. If there is any HF action going on, I prob. should add a 1nF ceramic disc?
Offline Gyro
« Reply #14 on: February 10, 2017, 09:05:36 pm »
I've just noticed that the manuals are now available for download...
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=06_Misc_Test_Equipment/Datron
In my own defense I should point out that I was working from a photocopy and don't have access to an A3 scanner (1041 Service manual). :D I did make sure that there is plenty of overlap on the schematic pages though.
The 1051 manual has much more detail than mine (another photocopy), so I might stand a better chance of getting my 1055 A/D board working again.
Best Regards, Chris
Offline Gyro
« Reply #15 on: February 10, 2017, 09:13:25 pm »
@gyro
Yes, unfortunately they are getting a lot less common. They do come up on ebay occasionally for little money (compared to the 106x plus). There was a flurry of 1041Ms in the UK a while back, obviously some establishment clearing out. Worth keeping an eye out for though.
Best Regards, Chris
« Reply #16 on: February 26, 2017, 07:02:30 pm »
Some progess...
I replaced all tantal caps. Some with decent eletrolytics, some with tantals with much higher rating (16V -> 35V oder 40V). Found another roasted one (C26), it is totally black colored (was as blue as the rest before) and had a little hole.. You can see it on the main board just above the two 1000µF electrolytics, right beneath a good blue one. Also a 16V rated that is connected to the 15V rail.
The unit powers up and I can measure DC voltage - but only up to 19.999mV. As soon as it gets to and above 20mV the displays is blank.  ??? For example +14.68mV is shown as +1.4667V - see picture. Have not tried AC yet, simply because I have nothing that can give me max. 20mV AC..
I can trim to 0.0000 with shorted leads with the potentiometer on the DC board. All range switches have no effect at all, neither Auto nor the voltage ranges do anything. DC/AC switch seems to work and I can see some floating AC noise on the DC supply I used.
Offline Gyro
« Reply #17 on: February 26, 2017, 09:33:19 pm »
Well that looks like significant progress.  :-+
It's encouraging that you have mV readings (even if it shows up on the wrong range) and the ability to zero the meter, that means that the critical DC gain block is working and probably undamaged.
You've probably realised by now but your problem lies somewhere in the front portion of the DC Isolator board - the small area between the push buttons and the  six optoisolators which drive the gain switching fets and relay. The first thing to check is the ribbon cable between the two boards - it could be something as simple as a bent pin preventing the 5V DC supply reaching the range logic.
It should be a fairly simple task to tie down the fault with reference to the DC Isolator board schematic. Note that the logic in the range selection area is 74C series, not TTL (HC would probably substitute ok as long as there are no timing issues). Make sure the remote button is out as it will otherwise disable the logic (check the signal from the switch in case of bad contacts). Concentrate on manual ranging, once that works you'll probably find that autorange starts working too. The range logic also controls decimal point position on the display and mV/V annunciators (via that same ribbon cable), - another useful debug indication (as it is proving to be already).
Best Regards, Chris
Offline eblc1388
« Reply #18 on: February 27, 2017, 12:54:40 am »
As per Gyro's advice, please check that the logic levels on these ICs output pins shown below do change when you manually selects different range settings or DVM functions. I have had a working 74C157 gone bad for no apparant reason.
« Reply #19 on: February 27, 2017, 04:01:53 am »
Yeah, already poked around in this area. Cable seems fine, all resistors show the correct values - besides two, gonna check them later. Transistor and diodes checked, reseated all ICs and checked the voltages - 4.6-4.8V, a bit low. When I disconnect the DC input board, the voltage on the main board goes up to 4.9V. The 7805 seems OK though a bit rotten. Nothing gets hot, no obvious damaged parts.
The last 3 wires (J4 14-16) transfer the data, next step will be hooking up my scope and check if I can see any signals and check the levels.
74C NOS are hard to get, prob. have some 74HCT.. hmm..
Quote from: eblc1388
Some kids help me repairing, so we will try to draw a decent table.. they need to learn something.. *g*
« Reply #20 on: February 27, 2017, 07:58:04 pm »
Funny meter. Today it only measures in the 1V-Rang ???, bang on with my Sanwa. But not above 1.9999 - as yesterday within the 10mV range, range switches still off duty. Seems a little bit random on what range the meter starts - and stays.
Switching area around M1 und M2 seem to be OK - on the input side. At least we see H and L when switching, we see Auto on/off etc.
We can see the corresponding H and L on the input of the 74C193, but not on the output.
M2/74C193:
Pin 1,10,11(Auto),15 - H or L when switching
2,6 - 0.9V
3 - 0.1V
4,5,12,16 - all H
7,8,9,14 - all ground/low
The output is 0.9V on pin 2 and 6 and 0.1V on pin 3 no matter what is on the input. Oh well... getting complicated..  :'(
Offline Kleinstein
« Reply #21 on: February 27, 2017, 08:34:49 pm »
The connection to pin 11 could be a problem -  so not loading new values to the C193.
The 74HC193 should be a suitable replacement. HCT might work too, but the input levels are slightly different.
With a rather week output of the 74C193 the 0.9 V level might be still a valid low with no real problem.
« Reply #22 on: February 27, 2017, 09:41:50 pm »
Pin 11 is either high or low if Auto button is pushed or not, so it works as it should I guess. But pin 2 and 6 always stay at ~0.9V and pin3 is always low at >0.1V.
If I short pin 1 and 2 = pull output 2 high, a relais klicks and the display switches from 1.9032V to 0.0019V - prob. 1kV range or 100V range.
Offline Gyro
« Reply #23 on: February 28, 2017, 08:06:11 pm »
It sounds as if it might be a faulty 74C193 (M2) then.  :-\ ...Although it would also be worth checking the outputs again with M7 (3 input NAND) removed in case it's inputs have failed (or its VCC missing) and loading down M2's outputs. Also check with M4 removed (2 input NAND).
Sorry to sound like a stuck record, but two things...
Firstly do leave Auto alone for now, in manual mode, then M2 is just operating in parallel input mode (pin 11, parallel load, low). rather than doing any counting - the outputs should just follow the inputs.
Secondly double check that the Remote signal to M1 pin 1 and M8 pin 1 (via D24). This selects between push button inputs and the remote control inputs (which would come in via J4, the unpopulated dip socket next to the ribbon if the option was fitted). There is a danger that you may be getting unpredictable data into M2 if the remote select signal is bad.
As I say, in manual mode, the outputs of M2 should be following it's inputs, as static signals. If this isn't happening then that's where you need to concentrate with a meter, continuity, shorts etc.
« Last Edit: February 28, 2017, 08:08:35 pm by Gyro »
Best Regards, Chris
« Reply #24 on: March 01, 2017, 02:50:34 am »
I already poked around that area behind M2, checked the opto couplers and transistors and also removed both NAND gates (M4 & M7) that are connected to the output of M2. No change, the 3 outputs stay at those levels. Will get replacement HC chips in a few days. Those 74-series chips are really funny - you always have a big pile of them - but never ever the one you need.. ^^
We only tested if Auto, DC, AC etc. switches actually do anything. i.e. pull some input high or low and checked the results with the schematics.
Well, the outputs of M1 and M8 change state when pressing the range switches, so M2 gets various inputs - but does not reflect the signals on the output side, they stay low or undefined permanently. So I don't think a wrong remote signal is involved.
I don't see any signals with higher frequencies around that range select section anyway, it's fairly just a bunch of static (logic) switches (at least in maual mode).
Checked all that could be involved (or better: I let it get checked by the kids  :popcorn:, that's how we noticed that shorting pin 1&2 on M2 activates the relay..  :-/O) This repair is a kind of extracurricular private activity project for some kids interested in electronics). We now wait for the spare parts and prob. build a little circuit in which we can check the 74C193 - or build some more cheap chinese DIY kits awhile.
The good thing on those older stuff is, that you can actually 1. repair them and 2. get the schematics and 3. at least have an idea how they work. Repairing an eg. iPhone or a Playstation is more or less.. just changing parts.
« Last Edit: March 01, 2017, 02:19:54 pm by Zeitkind »