Hello everybody, I'm struggling to repair a DATRON 1071 got from a rather well known German seller of used equipment, with the comment similar with "...I've got tired to replace the bloody tantalums, besides that, all is perfect...", yeah sure

.
OK, the display was nice and bright, the device arrived and besides a lot of dead capacitors, both tantalums and elkos, it was having the dreaded corrosion pattern, some chips were having some kind of talcum like dust on pins, the wires were strongly corroded (I had to replace the wires to the 5V regulator, among others) and the front plastic mask was having the glue dried and the orange filter curbed.
Alright, replaced the caps, redid the wiring (with similar teflon isolated wire), reglued the front panel mask, did some amateurish basic calibration and let it run to test the long term stability.
Now comes the strange part: after 8-10 hours, I hear some relay switching ca. 2 times/s !!!, everything seems to work OK, I open again the lids and is RL2 on the AC voltage board, run tests, test run OK, switch to AC, now TWO relays (one on the analogue board (the small one) are clicking, strangely the measurements seem to work OK, autotest procedure completes without errors !!!, just the bloody relay is clacking annoyingly.
I thought that the optocouplers went bad, I've measured them in and out (painful

) perfect defined signals, tried shuffling them, no change, WTF, look at the buffers and flipflops, nothing, all signals are perfect.
The issue is clear because of the digital command that comes during the acquisition procedure, if I put the device on hold and do manual sampling I got exactly the same behavior (one click-clack per conversion)
Yesterday the defect was gone when the device was cold, today is there from the beginning, I'm completely baffled, I was hitting the boards with a pencil, seem to not be a loose contact, what should I try next ?
Many thanks,
DC1MC