Author Topic: Datron 4805 multifunction Calibrator - No DC output voltage.  (Read 681 times)
Offline CGMtestingTopic starter
  • Country: mt
Dear All,
I am new to this forum and thanks for accepting me.
I have an issue on my Datron 4805. My problem is that for some reason, I have no DC voltage as output....The output is always 0Vdc.
Does anyone have any idea where to check because I have no schematics to start.
Thanks.
Online Grandchuck
  • Country: mt
Re: Datron 4805 multifunction Calibrator - No DC output voltage.
« Reply #2 on: May 08, 2020, 05:44:19 am »
Hi Grandchuck,
Thanks a lot for helping me.
I found the fault .....found 2 dry joints on Q6 (JE340) and Q7(JE350) of the Vdc card (with orange circuit board ejectors).  On the collector I had to find +38Vdc and -38Vdc respectively.
Thank you.
Online Grandchuck
« Reply #3 on: May 08, 2020, 02:06:11 pm »
Nice work!  You are welcome.