EEVblog Electronics Community Forum
Electronics => Repair => Topic started by: CGMtesting on May 07, 2020, 07:16:51 am
-
Dear All,
I am new to this forum and thanks for accepting me.
I have an issue on my Datron 4805. My problem is that for some reason, I have no DC voltage as output....The output is always 0Vdc.
Does anyone have any idea where to check because I have no schematics to start.
Thanks.
-
Here is a link for manuals for two similar models:
https://www.eevblog.com/forum/testgear/service-manuals-for-datronwavetek-4808-multifunction-calibrator/ (https://www.eevblog.com/forum/testgear/service-manuals-for-datronwavetek-4808-multifunction-calibrator/)
-
Hi Grandchuck,
Thanks a lot for helping me.
I found the fault .....found 2 dry joints on Q6 (JE340) and Q7(JE350) of the Vdc card (with orange circuit board ejectors). On the collector I had to find +38Vdc and -38Vdc respectively.
Thank you.
-
Nice work! You are welcome.