Author Topic: Datron 4920  (Read 363 times)
Offline NiaDebesis
Datron 4920
« on: April 10, 2021, 08:40:41 pm »
I have in my hands a good 4920 with full options, mv, DC, and other things.
Unfortunally it is broken, on the left display says "Error UR" with I didn't find in the manual and the right display is gone.
Thinking about a recapping and checking the PSU and the linear regulators.
But anyone know what Error UR means?
Best regards
Offline TheSteve
  • Living the Dream
Re: Datron 4920
« Reply #1 on: April 11, 2021, 04:22:08 am »
Error UR - is completely normal. It means "Error Under Range" - all ranges need a minimum input voltage to display a reading. It is documented in the manual.
VE7FM
Offline NiaDebesis
Re: Datron 4920
« Reply #2 on: April 11, 2021, 08:10:21 am »
Yes yesterday I found out this because the second display is dead.
I have lost cal data so it tells me to press some buttons that I didint know off.
For cal data it means only the calibration data of the procedure right? because I can recalibrate it
Offline picburner
Re: Datron 4920
« Reply #3 on: April 11, 2021, 02:12:53 pm »
If your right display is broken you can try my solution, I tried it with a Datron 1271 but it should work on your Datron 4920 too.
Offline TiN
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Datron 4920
« Reply #4 on: April 13, 2021, 02:46:41 am »
Recalibration of 4920 is lot of fun, unless you work in high-end metrology lab  :D
YouTube | Metrology IRC Chat room | Live-cam | Share T&M documentation? Upload! No MB limit, firmwares, photos.