Hi,

I have in my hands a good 4920 with full options, mv, DC, and other things.
Unfortunally it is broken, on the left display says "Error UR" with I didn't find in the manual and the right display is gone.
Thinking about a recapping and checking the PSU and the linear regulators.

But anyone know what Error UR means?


Best regards
 

Online TheSteve

  • Supporter
  • ****
  • Posts: 3753
  • Country: ca
  • Living the Dream
Re: Datron 4920
« Reply #1 on: April 11, 2021, 04:22:08 am »
Error UR - is completely normal. It means "Error Under Range" - all ranges need a minimum input voltage to display a reading. It is documented in the manual.
VE7FM
 

Offline NiaDebesisTopic starter

  • Frequent Contributor
  • **
  • Posts: 278
  • Country: it
Re: Datron 4920
« Reply #2 on: April 11, 2021, 08:10:21 am »
Yes yesterday I found out this because the second display is dead.
I have lost cal data so it tells me to press some buttons that I didint know off.

For cal data it means only the calibration data of the procedure right? because I can recalibrate it
 

Offline picburner

  • Frequent Contributor
  • **
  • Posts: 498
  • Country: it
Re: Datron 4920
« Reply #3 on: April 11, 2021, 02:12:53 pm »
If your right display is broken you can try my solution, I tried it with a Datron 1271 but it should work on your Datron 4920 too.
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Datron 4920
« Reply #4 on: April 13, 2021, 02:46:41 am »
Recalibration of 4920 is lot of fun, unless you work in high-end metrology lab  :D
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline chinapp

  • Contributor
  • Posts: 33
  • Country: 00
Re: Datron 4920
« Reply #5 on: March 06, 2022, 03:05:52 pm »
could you share more pics inside ? I wanna see what's inside the mV option