Twas supposed to be a simple day

. Check and Calibrate a BK 1602 Valve power supply (with the keithley 2110 ) to sell.. Got the schematics and the procedure (just twiddling some pots) .

I got a whiff off something heating up in the power supply when setting a meter cal ,and I thought "Hey ,The 2110 has a temperature second display while measuring volts.. So lets turn that on ,and touch this transistor ... " and Zap!!!

?? The temp input has opto isolation and should be isolated from the Main measurement side ..
Now, The keithley 2110 is going nuts on the display with no input.. and lets look at the BK 1602 .. won't reach full 400V without the current limit tripping ..(with no load!!) .. Kill two units with one thermocouple..

Im wary of ground short circuits via probes after I watched Dave's "Dont blow the ass out of your scope ...video" way back , and In this case my beliefs were that the thermo inputs were isolated ..
The aftermath .. I didn't realize this at first but there was probably a mini fire in the scope when it was letting its magic smoke out .... and the smell .. OMG , phenol ,phenols all the way ..
And then I opened the unit (smell to infinity ) ... and prominently left the space.. Two days of airing out the room and running a mini ozoniser later .. lol
So , it looks like an LM393 is visibly dead and maybe others bits around it . Looking at forums it looks like another member (with a 2700) had one go out as well .
Questions
The pads have delaminated and I have probed the circuit for connections and Annotated them.
I have been slowly scraping away the blackened section ,but it looks like I may have to clear out the traces and leave the area unpopulated.
I cant cut a hole out in the pcb because there are traces on the bottom . There may be hidden traces as well .
Since the pcb is charred and shows a 6 M resistance on the fiberglass is it worth reusing the area for new parts ? ,or do a riser / daughterboard away from the burn spot ?
If I can relocate the circuit ,should I dremel out as much as possible to prevent leakages etc?The top phenolic layer is gone and Im seeing fiberglass mat .
Im thinking this is the start of problems and more will reveal themselves (whack a mole ,lol)