DMM failing to meter DC voltage unless after minute waiting
« on: November 07, 2022, 01:27:05 am »
Anyone experienced how to solve a DMM (of medium quality not top/best one) failing to meter DC voltage unless after probing several times later or perhaps just waiting a minute or so later, after which it will work normally for the needed measurement
 

Offline BILLPOD

  • Frequent Contributor
  • **
  • Posts: 361
  • Country: us
Re: DMM failing to meter DC voltage unless after minute waiting
« Reply #1 on: November 07, 2022, 07:36:14 pm »
abdulbadii, What is the make and model of your medium quality DMM?
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: DMM failing to meter DC voltage unless after minute waiting
« Reply #2 on: November 07, 2022, 08:17:37 pm »
Sounds like a bad contact or open (fusible?) resistor in series with the input to me. What is the input that you measure when you measure resistance with the meter set to DCV between the V/Ohm terminaland COM terminal with another DMM, and what is the input impedance specified in the manual for that DCV range? And is this value stable or does it slowly change like the reading on the defective DMM?
« Last Edit: November 07, 2022, 08:23:34 pm by alm »
 
The following users thanked this post: abdulbadii

Offline Grandchuck

  • Frequent Contributor
  • **
  • Posts: 725
  • Country: us
Re: DMM failing to meter DC voltage unless after minute waiting
« Reply #3 on: November 07, 2022, 09:25:39 pm »
Are you sure it is the meter?  Do you have another one?
 