| charlyd:
Hi i got a DPO7254 on my bench, after starting up the unit, it fails with error 315 and 316. My debug logs are pointing to a demux communication problem. my error looks like this: --- Code: ---*************************************** File opened: Sat Jan 07 13:52:30 2023 *************************************** EnablePowerupDiags = 1, puMode=User DiagPowerUpDiagsEnabled = 1 Model: DPO7254 Serial number: B010315, HFD201[0] ID = 0 (0x000000) Scope.exe Version: 5.3.5 Build 22; Sat Jan 07 13:52:41 2023: Executing Powerup diags ... Pu=24 T=27C Test "111" - Processor-->Memory-->DRAMWalk1 started Test execution time 9.7778e-006 sec Test "112" - Processor-->Memory-->DRAMCell started Test execution time 0.010838 sec Test "113" - Processor-->Memory-->DRAMMarch started Test execution time 0.0015251 sec Test "114" - Processor-->Memory-->DRAMCal started Test execution time 2.0394e-005 sec Test "211" - Registers-->PCI-->MIA started Test execution time 0.0056395 sec Test "221" - Registers-->Acquisition-->ACL started Test execution time 0.00019779 sec Test "222" - Registers-->Acquisition-->PLL started Test execution time 0.020034 sec Test "223" - Registers-->Acquisition-->Preamp started Test execution time 0.0067425 sec Test "224" - Registers-->Acquisition-->TrkHld started Test execution time 0.0040994 sec Test "225" - Registers-->Acquisition-->ADC started Test execution time 0.023456 sec Test "226" - Registers-->Acquisition-->DAC started Test execution time 0.005549 sec Test "227" - Registers-->Acquisition-->Demux started Test execution time 0.0040382 sec Test "231" - Registers-->Trigger-->SBTL started Test execution time 8.7721e-005 sec Test "232" - Registers-->Trigger-->SGTL started Test execution time 0.0067855 sec Test "233" - Registers-->Trigger-->TrigComp started Test execution time 9.3587e-005 sec Test "234" - Registers-->Trigger-->CommTrig started Test execution time 0.034591 sec Test "235" - Registers-->Trigger-->FTL started *** TEST N/A FOR THIS HW *** Test execution time 1.9556e-006 sec Test "241" - Registers-->Misc-->FanCtrl started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "242" - Registers-->Misc-->Tek04 started *** TEST N/A FOR THIS HW *** Test execution time 1.9556e-006 sec Test "243" - Registers-->Misc-->DigCtrl started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "244" - Registers-->Misc-->IntfBoard started *** TEST N/A FOR THIS HW *** Test execution time 1.9556e-006 sec Test "245" - Registers-->Misc-->FpBoard started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "246" - Registers-->Misc-->AfeBoard started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "252" - Registers-->DigitalTrig-->Tek05Dev started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "253" - Registers-->DigitalTrig-->Tek05HKRam started *** TEST N/A FOR THIS HW *** Test execution time 1.9556e-006 sec Test "254" - Registers-->DigitalTrig-->Tek05AcqRam started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "311" - Acquisition-->Demux-->RunAB started Test execution time 0.28398 sec Test "312" - Acquisition-->Demux-->AcqDone started Test execution time 0.0016488 sec Test "313" - Acquisition-->Demux-->SysRdy started Test execution time 0.0015136 sec Test "314" - Acquisition-->Demux-->Interrupt started Test execution time 0.0014753 sec Test "315" - Acquisition-->Demux-->IdcLoSpeed started Error from DEMUX3 to DEMUX4 expect: 000037fe rcv: 00003f56 Error from DEMUX3 to DEMUX4 expect: 000037fd rcv: 00003f55 Error from DEMUX3 to DEMUX4 expect: 000037fb rcv: 00003f53 Error from DEMUX3 to DEMUX4 expect: 000037f7 rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000037ef rcv: 00003f47 Error from DEMUX3 to DEMUX4 expect: 000037df rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000037bf rcv: 00003f17 Error from DEMUX3 to DEMUX4 expect: 0000377f rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000036ff rcv: 00003e57 Error from DEMUX3 to DEMUX4 expect: 000035ff rcv: 00003d57 Error from DEMUX3 to DEMUX4 expect: 000033ff rcv: 00003b57 Error from DEMUX3 to DEMUX4 expect: 00003fff rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000027ff rcv: 00002f57 Error from DEMUX3 to DEMUX4 expect: 000017ff rcv: 00001f57 Error from DEMUX3 to DEMUX4 expect: 000037ff rcv: 00003f57 Test diagIdcLoSpeedRunTest TransmitRight - WalkZeros FAILED Error from DEMUX3 to DEMUX4 expect: 00000808 rcv: 00000800 Error from DEMUX3 to DEMUX4 expect: 00000820 rcv: 00000800 Error from DEMUX3 to DEMUX4 expect: 00000880 rcv: 00000800 Error from DEMUX3 to DEMUX4 expect: 00000000 rcv: 00000800 Test diagIdcLoSpeedRunTest TransmitRight - WalkOnes FAILED Error from DEMUX4 to DEMUX3 expect: 000037fe rcv: 00003f56 Error from DEMUX4 to DEMUX3 expect: 000037fd rcv: 00003f55 Error from DEMUX4 to DEMUX3 expect: 000037fb rcv: 00003f53 Error from DEMUX4 to DEMUX3 expect: 000037f7 rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000037ef rcv: 00003f47 Error from DEMUX4 to DEMUX3 expect: 000037df rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000037bf rcv: 00003f17 Error from DEMUX4 to DEMUX3 expect: 0000377f rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000036ff rcv: 00003e57 Error from DEMUX4 to DEMUX3 expect: 000035ff rcv: 00003d57 Error from DEMUX4 to DEMUX3 expect: 000033ff rcv: 00003b57 Error from DEMUX4 to DEMUX3 expect: 00003fff rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000027ff rcv: 00002f57 Error from DEMUX4 to DEMUX3 expect: 000017ff rcv: 00001f57 Error from DEMUX4 to DEMUX3 expect: 000037ff rcv: 00003f57 Test diagIdcLoSpeedRunTest TransmitLeft - WalkZeros FAILED Error from DEMUX4 to DEMUX3 expect: 00000808 rcv: 00000800 Error from DEMUX4 to DEMUX3 expect: 00000820 rcv: 00000800 Error from DEMUX4 to DEMUX3 expect: 00000880 rcv: 00000800 Error from DEMUX4 to DEMUX3 expect: 00000000 rcv: 00000800 Test diagIdcLoSpeedRunTest TransmitLeft - WalkOnes FAILED IDC Low Speed test FAILED diagIdcLoSpeed error summary of failing demux pairs: -------------------------------------------------- DEMUX3 --> DEMUX4 DEMUX4 --> DEMUX3 *** TEST FAILED *** Test execution time 0.020886 sec diagMgr:diagLogFailure: Sat Jan 07 13:52:41 2023 P Pu=24 T=27 "diagIdcLoSpeed FAILED" Test "316" - Acquisition-->Demux-->IdcHiSpeed started Error from DEMUX3 to DEMUX4 expect: ffff7ffe rcv: ffff7f7e Error from DEMUX3 to DEMUX4 expect: fffefffd rcv: ff7eff7d Error from DEMUX3 to DEMUX4 expect: fff7ffef rcv: ff77ff6f Error from DEMUX3 to DEMUX4 expect: ffdfffbf rcv: ff5fff3f Error from DEMUX3 to DEMUX4 expect: 00400080 rcv: 00400000 Error from DEMUX3 to DEMUX4 expect: fefffdff rcv: fe7ffd7f Error from DEMUX3 to DEMUX4 expect: f7ffefff rcv: 77ffefff Error from DEMUX3 to DEMUX4 expect: dfffbfff rcv: 5fffbfff Error from DEMUX3 to DEMUX4 expect: bfff7fff rcv: bfff7f7f Test FAILED Error from DEMUX4 to DEMUX3 expect: ffff7ffe rcv: ffff7f7e Error from DEMUX4 to DEMUX3 expect: fffefffd rcv: 7f7e7f7d Error from DEMUX4 to DEMUX3 expect: fffdfffb rcv: 7f7d7f7b Error from DEMUX4 to DEMUX3 expect: fffbfff7 rcv: 7f7b7f77 Error from DEMUX4 to DEMUX3 expect: fff7ffef rcv: 7f777f6f Error from DEMUX4 to DEMUX3 expect: ffefffdf rcv: 7f6f7f5f Error from DEMUX4 to DEMUX3 expect: ffdfffbf rcv: 7f5f7f3f Error from DEMUX4 to DEMUX3 expect: 00400080 rcv: 00400000 Error from DEMUX4 to DEMUX3 expect: ff7ffeff rcv: ff7f7e7f Error from DEMUX4 to DEMUX3 expect: fefffdff rcv: 7e7f7d7f Error from DEMUX4 to DEMUX3 expect: fdfffbff rcv: 7d7f7b7f Error from DEMUX4 to DEMUX3 expect: fbfff7ff rcv: 7b7f777f Error from DEMUX4 to DEMUX3 expect: f7ffefff rcv: 777f6f7f Error from DEMUX4 to DEMUX3 expect: efffdfff rcv: 6f7f5f7f Error from DEMUX4 to DEMUX3 expect: dfffbfff rcv: 5f7f3f7f Error from DEMUX4 to DEMUX3 expect: bfff7fff rcv: bfff7f7f Test FAILED diagIdcHiSpeed error summary of failing demux pairs: -------------------------------------------------- DEMUX3 --> DEMUX4 DEMUX4 --> DEMUX3 Retrying tstDiagIdcHiSpeed test ... T=27C Error from DEMUX3 to DEMUX4 expect: fffefffd rcv: ff7eff7d Error from DEMUX3 to DEMUX4 expect: fff7ffef rcv: ff77ffef Error from DEMUX3 to DEMUX4 expect: ffdfffbf rcv: ff5fff3f Error from DEMUX3 to DEMUX4 expect: 00400080 rcv: 00400000 Error from DEMUX3 to DEMUX4 expect: fdfffbff rcv: fd7ffbff Error from DEMUX3 to DEMUX4 expect: f7ffefff rcv: f77fef7f Test FAILED Error from DEMUX4 to DEMUX3 expect: ffff7ffe rcv: ffff7f7e Error from DEMUX4 to DEMUX3 expect: fffefffd rcv: 7f7e7f7d Error from DEMUX4 to DEMUX3 expect: fffdfffb rcv: 7f7d7f7b Error from DEMUX4 to DEMUX3 expect: fffbfff7 rcv: 7f7b7f77 Error from DEMUX4 to DEMUX3 expect: fff7ffef rcv: 7f777f6f Error from DEMUX4 to DEMUX3 expect: ffefffdf rcv: 7f6f7f5f Error from DEMUX4 to DEMUX3 expect: ffdfffbf rcv: 7f5f7f3f Error from DEMUX4 to DEMUX3 expect: 00400080 rcv: 00400000 Error from DEMUX4 to DEMUX3 expect: ff7ffeff rcv: ff7f7e7f Error from DEMUX4 to DEMUX3 expect: fefffdff rcv: 7e7f7d7f Error from DEMUX4 to DEMUX3 expect: fdfffbff rcv: 7d7f7b7f Error from DEMUX4 to DEMUX3 expect: fbfff7ff rcv: 7b7f777f Error from DEMUX4 to DEMUX3 expect: f7ffefff rcv: 777f6f7f Error from DEMUX4 to DEMUX3 expect: efffdfff rcv: 6f7f5f7f Error from DEMUX4 to DEMUX3 expect: dfffbfff rcv: 5f7f3f7f Error from DEMUX4 to DEMUX3 expect: bfff7fff rcv: bfff7f7f Test FAILED diagIdcHiSpeed error summary of failing demux pairs: -------------------------------------------------- DEMUX3 --> DEMUX4 DEMUX4 --> DEMUX3 *** TEST FAILED *** Test execution time 0.56149 sec diagMgr:diagLogFailure: Sat Jan 07 13:52:41 2023 P Pu=24 T=27 "diagIdcHiSpeed FAILED" Test "317" - Acquisition-->Demux-->IdcAcqDataXfr started Test skipped because Demux IDC tests failed. *** TEST NOT RUN YET *** Test execution time 0.0001841 sec Test "318" - Acquisition-->Demux-->IdcDispDataXfr started Test skipped because Demux IDC tests failed. *** TEST NOT RUN YET *** Test execution time 0.00015449 sec Test "321" - Acquisition-->Dma-->PaDma started Test skipped because Demux IDC tests failed. *** TEST NOT RUN YET *** Test execution time 0.00017349 sec Test "322" - Acquisition-->Dma-->IdcAcqDma started Test skipped because Demux IDC tests failed. *** TEST NOT RUN YET *** Test execution time 0.00035591 sec Test "331" - Acquisition-->Memory-->MemData started Test execution time 0.020599 sec Test "332" - Acquisition-->Memory-->MemAddr started Test execution time 0.0038114 sec Test "342" - Acquisition-->Vertical-->TH Inputs started Test execution time 0.30105 sec Test "345" - Acquisition-->Vertical-->50ohmOvld started Test execution time 0.90188 sec Test "351" - Acquisition-->PLL-->Clock Freq started Test execution time 0.65833 sec Test "371" - Acquisition-->AcqProcessor-->SPI started Test execution time 0.00016874 sec Test "372" - Acquisition-->AcqProcessor-->INT Line started Test execution time 0.00069087 sec Test "373" - Acquisition-->AcqProcessor-->Interrupt started Test execution time 0.00099873 sec Test "374" - Acquisition-->AcqProcessor-->RelayDrive started *** TEST N/A FOR THIS HW *** Test execution time 3.073e-006 sec Test "375" - Acquisition-->AcqProcessor-->TWI started Test execution time 0.00013102 sec Test "376" - Acquisition-->AcqProcessor-->PrbInterrupts started Test execution time 0.00036317 sec Test "377" - Acquisition-->AcqProcessor-->TCpowerCtrl started *** TEST N/A FOR THIS HW *** Test execution time 1.6762e-006 sec Test "378" - Acquisition-->AcqProcessor-->MAX517 started Test execution time 0.00093978 sec Test "411" - Trigger-->Inputs-->Ch1 started Test execution time 0.14072 sec Test "412" - Trigger-->Inputs-->Ch2 started Test execution time 0.13072 sec Test "413" - Trigger-->Inputs-->Ch3 started Test execution time 0.13071 sec Test "414" - Trigger-->Inputs-->Ch4 started Test execution time 0.1307 sec Test "415" - Trigger-->Inputs-->Line started Test execution time 0.014745 sec Test "416" - Trigger-->Inputs-->Video started Test execution time 0.21615 sec Test "417" - Trigger-->Inputs-->Events started Test execution time 0.22039 sec Test "418" - Trigger-->Inputs-->Serial started Test execution time 0.3515 sec Test "419" - Trigger-->Inputs-->HSMeta started *** TEST N/A FOR THIS HW *** Test execution time 2.2349e-006 sec Test "421" - Trigger-->Outputs-->523TrgOut started Test execution time 0.00050565 sec Test "422" - Trigger-->Outputs-->523CpuInt started Test execution time 0.00047324 sec Test "423" - Trigger-->Outputs-->Fedge started Test execution time 0.043754 sec Test "424" - Trigger-->Outputs-->Ftrig started Test execution time 0.043379 sec Test "441" - Trigger-->Word-->Compare started *** TEST N/A FOR THIS HW *** Test execution time 2.5143e-006 sec Test "442" - Trigger-->Word-->Delay started *** TEST N/A FOR THIS HW *** Test execution time 2.2349e-006 sec Test "443" - Trigger-->Word-->Clk Qualify started *** TEST N/A FOR THIS HW *** Test execution time 1.9556e-006 sec Test "444" - Trigger-->Word-->Formed LSST started *** TEST N/A FOR THIS HW *** Test execution time 2.2349e-006 sec Test "451" - Trigger-->Digital-->Acq started *** TEST N/A FOR THIS HW *** Test execution time 1.9556e-006 sec Sat Jan 07 13:52:46 2023: USER MODE POWERUP DIAGNOSTICS RESULTS: 22 (of 67) tests SKIPPED Sat Jan 07 13:52:46 2023: USER MODE POWERUP DIAGNOSTICS RESULTS: 2 (of 45) EXECUTED tests FAILED Processor Failures: 0 Registers Failures: 0 Acquisition Failures: 2 Trigger Failures: 0 TekLink Failures: 0 Test 315 - F ==> Acquisition->Demux->IdcLoSpeed, 0.020886 sec diagIdcLoSpeed FAILED Test 316 - F ==> Acquisition->Demux->IdcHiSpeed, 0.56149 sec diagIdcHiSpeed FAILED 13:52:46:048[2336] diagMgr.cpp:calDiagManager(948) Powerup diags finished in 4.9161 sec << POWERUP DIAGS FINISHED >> Sat Jan 07 14:10:05 2023: DIAGNOSTIC EXECUTION starting ... Pu=24 T=31C Test "311" - Acquisition-->Demux-->RunAB started Test execution time 0.30136 sec Test "312" - Acquisition-->Demux-->AcqDone started Test execution time 0.0016745 sec Test "313" - Acquisition-->Demux-->SysRdy started Test execution time 0.0012745 sec Test "314" - Acquisition-->Demux-->Interrupt started Test execution time 0.0031004 sec Test "315" - Acquisition-->Demux-->IdcLoSpeed started Error from DEMUX3 to DEMUX4 expect: 000037fe rcv: 00003f56 Error from DEMUX3 to DEMUX4 expect: 000037fd rcv: 00003f55 Error from DEMUX3 to DEMUX4 expect: 000037fb rcv: 00003f53 Error from DEMUX3 to DEMUX4 expect: 000037f7 rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000037ef rcv: 00003f47 Error from DEMUX3 to DEMUX4 expect: 000037df rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000037bf rcv: 00003f17 Error from DEMUX3 to DEMUX4 expect: 0000377f rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000036ff rcv: 00003e57 Error from DEMUX3 to DEMUX4 expect: 000035ff rcv: 00003d57 Error from DEMUX3 to DEMUX4 expect: 000033ff rcv: 00003b57 Error from DEMUX3 to DEMUX4 expect: 00003fff rcv: 00003f57 Error from DEMUX3 to DEMUX4 expect: 000027ff rcv: 00002f57 Error from DEMUX3 to DEMUX4 expect: 000017ff rcv: 00001f57 Error from DEMUX3 to DEMUX4 expect: 000037ff rcv: 00003f57 Test diagIdcLoSpeedRunTest TransmitRight - WalkZeros FAILED Error from DEMUX3 to DEMUX4 expect: 00000808 rcv: 00000800 Error from DEMUX3 to DEMUX4 expect: 00000820 rcv: 00000800 Error from DEMUX3 to DEMUX4 expect: 00000880 rcv: 00000800 Error from DEMUX3 to DEMUX4 expect: 00000000 rcv: 00000800 Test diagIdcLoSpeedRunTest TransmitRight - WalkOnes FAILED Error from DEMUX4 to DEMUX3 expect: 000037fe rcv: 00003f56 Error from DEMUX4 to DEMUX3 expect: 000037fd rcv: 00003f55 Error from DEMUX4 to DEMUX3 expect: 000037fb rcv: 00003f53 Error from DEMUX4 to DEMUX3 expect: 000037f7 rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000037ef rcv: 00003f47 Error from DEMUX4 to DEMUX3 expect: 000037df rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000037bf rcv: 00003f17 Error from DEMUX4 to DEMUX3 expect: 0000377f rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000036ff rcv: 00003e57 Error from DEMUX4 to DEMUX3 expect: 000035ff rcv: 00003d57 Error from DEMUX4 to DEMUX3 expect: 000033ff rcv: 00003b57 Error from DEMUX4 to DEMUX3 expect: 00003fff rcv: 00003f57 Error from DEMUX4 to DEMUX3 expect: 000027ff rcv: 00002f57 Error from DEMUX4 to DEMUX3 expect: 000017ff rcv: 00001f57 Error from DEMUX4 to DEMUX3 expect: 000037ff rcv: 00003f57 Test diagIdcLoSpeedRunTest TransmitLeft - WalkZeros FAILED Error from DEMUX4 to DEMUX3 expect: 00000808 rcv: 00000800 Error from DEMUX4 to DEMUX3 expect: 00000820 rcv: 00000800 Error from DEMUX4 to DEMUX3 expect: 00000880 rcv: 00000800 Error from DEMUX4 to DEMUX3 expect: 00000000 rcv: 00000800 Test diagIdcLoSpeedRunTest TransmitLeft - WalkOnes FAILED IDC Low Speed test FAILED diagIdcLoSpeed error summary of failing demux pairs: -------------------------------------------------- DEMUX3 --> DEMUX4 DEMUX4 --> DEMUX3 *** TEST FAILED *** Test execution time 0.013694 sec diagMgr:diagLogFailure: Sat Jan 07 14:10:05 2023 D Pu=24 T=31 "diagIdcLoSpeed FAILED" Test "316" - Acquisition-->Demux-->IdcHiSpeed started Error from DEMUX3 to DEMUX4 expect: 00008001 rcv: 2828a829 Error from DEMUX3 to DEMUX4 expect: 00010002 rcv: 2829282a Error from DEMUX3 to DEMUX4 expect: 00020004 rcv: 282a282c Error from DEMUX3 to DEMUX4 expect: 00040008 rcv: 20242028 Error from DEMUX3 to DEMUX4 expect: fffbfff7 rcv: fffbffff Error from DEMUX3 to DEMUX4 expect: 00080010 rcv: 20282838 Error from DEMUX3 to DEMUX4 expect: 00100020 rcv: 08180828 Error from DEMUX3 to DEMUX4 expect: ffefffdf rcv: ffefffff Error from DEMUX3 to DEMUX4 expect: 00200040 rcv: 08282868 Error from DEMUX3 to DEMUX4 expect: 00400080 rcv: 286828a8 Error from DEMUX3 to DEMUX4 expect: 00800100 rcv: 28a82928 Error from DEMUX3 to DEMUX4 expect: 01000200 rcv: 29282a28 Error from DEMUX3 to DEMUX4 expect: 02000400 rcv: 2a282c28 Error from DEMUX3 to DEMUX4 expect: 04000800 rcv: 24202828 Error from DEMUX3 to DEMUX4 expect: 08001000 rcv: 28283828 Error from DEMUX3 to DEMUX4 expect: 10002000 rcv: 18082828 Error from DEMUX3 to DEMUX4 expect: 20004000 rcv: 28286828 Error from DEMUX3 to DEMUX4 expect: 40008000 rcv: 6828a828 Test FAILED Error from DEMUX4 to DEMUX3 expect: ffff7ffe rcv: ffff7f7e Error from DEMUX4 to DEMUX3 expect: fffefffd rcv: 7f7e7f7d Error from DEMUX4 to DEMUX3 expect: fffdfffb rcv: 7f7d7f7b Error from DEMUX4 to DEMUX3 expect: fffbfff7 rcv: 7f7b7f77 Error from DEMUX4 to DEMUX3 expect: fff7ffef rcv: 7f777f6f Error from DEMUX4 to DEMUX3 expect: ffefffdf rcv: 7f6f7f5f Error from DEMUX4 to DEMUX3 expect: ffdfffbf rcv: 7f5f7f3f Error from DEMUX4 to DEMUX3 expect: 00400080 rcv: 00400000 Error from DEMUX4 to DEMUX3 expect: ff7ffeff rcv: ff7f7e7f Error from DEMUX4 to DEMUX3 expect: fefffdff rcv: 7e7f7d7f Error from DEMUX4 to DEMUX3 expect: fdfffbff rcv: 7d7f7b7f Error from DEMUX4 to DEMUX3 expect: fbfff7ff rcv: 7b7f777f Error from DEMUX4 to DEMUX3 expect: f7ffefff rcv: 777f6f7f Error from DEMUX4 to DEMUX3 expect: efffdfff rcv: 6f7f5f7f Error from DEMUX4 to DEMUX3 expect: dfffbfff rcv: 5f7f3f7f Error from DEMUX4 to DEMUX3 expect: bfff7fff rcv: bfff7f7f Test FAILED diagIdcHiSpeed error summary of failing demux pairs: -------------------------------------------------- DEMUX3 --> DEMUX4 DEMUX4 --> DEMUX3 Retrying tstDiagIdcHiSpeed test ... T=31C Error from DEMUX3 to DEMUX4 expect: 00008001 rcv: 2828a829 Error from DEMUX3 to DEMUX4 expect: 00010002 rcv: 2829282a Error from DEMUX3 to DEMUX4 expect: 00020004 rcv: 282a282c Error from DEMUX3 to DEMUX4 expect: 00040008 rcv: 20242028 Error from DEMUX3 to DEMUX4 expect: fffbfff7 rcv: fffbffff Error from DEMUX3 to DEMUX4 expect: 00080010 rcv: 20282838 Error from DEMUX3 to DEMUX4 expect: 00100020 rcv: 08180828 Error from DEMUX3 to DEMUX4 expect: ffefffdf rcv: ffefffff Error from DEMUX3 to DEMUX4 expect: 00200040 rcv: 08282868 Error from DEMUX3 to DEMUX4 expect: 00400080 rcv: 286828a8 Error from DEMUX3 to DEMUX4 expect: 00800100 rcv: 28a82928 Error from DEMUX3 to DEMUX4 expect: 01000200 rcv: 29282a28 Error from DEMUX3 to DEMUX4 expect: 02000400 rcv: 2a282c28 Error from DEMUX3 to DEMUX4 expect: 04000800 rcv: 24202828 Error from DEMUX3 to DEMUX4 expect: 08001000 rcv: 28283828 Error from DEMUX3 to DEMUX4 expect: 10002000 rcv: 18082828 Error from DEMUX3 to DEMUX4 expect: 20004000 rcv: 28286828 Error from DEMUX3 to DEMUX4 expect: 40008000 rcv: 6828a828 Test FAILED Error from DEMUX4 to DEMUX3 expect: ffff7ffe rcv: ffff7f7e Error from DEMUX4 to DEMUX3 expect: fffefffd rcv: 7f7e7f7d Error from DEMUX4 to DEMUX3 expect: fffdfffb rcv: 7f7d7f7b Error from DEMUX4 to DEMUX3 expect: fffbfff7 rcv: 7f7b7f77 Error from DEMUX4 to DEMUX3 expect: fff7ffef rcv: 7f777f6f Error from DEMUX4 to DEMUX3 expect: ffefffdf rcv: 7f6f7f5f Error from DEMUX4 to DEMUX3 expect: ffdfffbf rcv: 7f5f7f3f Error from DEMUX4 to DEMUX3 expect: 00400080 rcv: 00400000 Error from DEMUX4 to DEMUX3 expect: ff7ffeff rcv: ff7f7e7f Error from DEMUX4 to DEMUX3 expect: fefffdff rcv: 7e7f7d7f Error from DEMUX4 to DEMUX3 expect: fdfffbff rcv: 7d7f7b7f Error from DEMUX4 to DEMUX3 expect: fbfff7ff rcv: 7b7f777f Error from DEMUX4 to DEMUX3 expect: f7ffefff rcv: 777f6f7f Error from DEMUX4 to DEMUX3 expect: efffdfff rcv: 6f7f5f7f Error from DEMUX4 to DEMUX3 expect: dfffbfff rcv: 5f7f3f7f Error from DEMUX4 to DEMUX3 expect: bfff7fff rcv: bfff7f7f Test FAILED diagIdcHiSpeed error summary of failing demux pairs: -------------------------------------------------- DEMUX3 --> DEMUX4 DEMUX4 --> DEMUX3 *** TEST FAILED *** Test execution time 0.56268 sec diagMgr:diagLogFailure: Sat Jan 07 14:10:05 2023 D Pu=24 T=31 "diagIdcHiSpeed FAILED" Test "317" - Acquisition-->Demux-->IdcAcqDataXfr started Test skipped because Demux IDC tests failed. *** TEST NOT RUN YET *** Test execution time 0.00020058 sec Test "318" - Acquisition-->Demux-->IdcDispDataXfr started Test skipped because Demux IDC tests failed. *** TEST NOT RUN YET *** Test execution time 0.00024277 sec Sat Jan 07 14:10:06 2023: DIAG MODE DIAGNOSTICS RESULTS: 2 (of 8) tests SKIPPED Sat Jan 07 14:10:06 2023: DIAG MODE DIAGNOSTICS RESULTS: 2 (of 6) EXECUTED tests FAILED Test 315 - F ==> Acquisition->Demux->IdcLoSpeed, 0.013694 sec diagIdcLoSpeed FAILED Test 316 - F ==> Acquisition->Demux->IdcHiSpeed, 0.56268 sec diagIdcHiSpeed FAILED DIAGNOSTIC EXECUTION finished in 0.94508 sec --- End code --- anybody have any experience in reflowing / reballing these guys without damaging the board or tracks below? |
| mikehank:
I have lots of experience reballing/replacing Demux's on these boards. I had access to a BGA rework machine. I'm not sure I would try a reflow on a lead free board. To hot and and risky. |
| charlyd:
today i took the chip off one of the two chips, i just warmed up the heatsink for 1,5 min at just over 400c. and then i could easy take it off the board. i found this on the bottom of the chip a little swallen... very little. ad a close up picture. but how can i find out the swallen part is still ok? diode test between the pads ? if somebody have a Donor board for sale from a DPO7000 series please get in touch with me. (also for just one chip contact me) National Semiconductor: ----------------------Demultiplexer---- VT626ES ADG387C MM9617C 167-0138-00 4673-0 |
| mikehank:
Here is what i would do. Your looking for open pads. Find a grounding pad(should be common with a bunch of pads). Then do a resistance check with pads in the bubbled area. One more thing to consider this bubble may not allow the BGA to sit correctly down. When I have manually pulled these chips is removed the heatsink 1st. Put a small thermocouple under the BGA(it will fit) and keep an eye on the temps as you heat the chip. |
| charlyd:
thanks for your reply i will certain check out the pads and see if it is still worthit. i got myself a rework station. i found a picture on the web of a simulair chip i am looking for. but what is the partnumber ? ADG387C or MM9617C-FBB or ? ( i am a little confused :-// ) who can help me out with a donor chip or board? a internet picture is shown below |