Then you haven't read enough on the forum. This has been posted a number of times. The a/d converter is broken. This is a common problem for these supplies.
You cannot measure in the a/d converter as it is a charge balancer. Everything happens in current domain.
The easiest way to repair is to swap the four ic's that make up the converter. There is an ad706 an ad711 , and two others. Look at the schematic of the a/d comverter. It is kade from an opamp a comparator and there is a vref generator making +10 and -10 volt with two opamps. Those are the bad boys.
Also ,look around the 4051 . The measuring signals as well as the vref is multiplexed. That multiplexr need replacing. Also measure the series resistors coming from vref. If they are the brownish colored ones with yellow lettering : i have seen those go bad...
Once that stuff is swapped you will need to calibrate the supply. Activate the self calibration the password is 003631.
You will need a voltmeter, a currentmeter and a load resistor.