Hi,
After waiting a week for precision resistors, today I got information that they are not even shipped out.

To get this process running again I got 1% SMD resistors and soldered them in series.
Results:
2M
2W 4W Fluke
002.092M 001.996M 1.995M
02.0751M 01.9958M
2.07507M 1.99575M
1M
2W 4W Fluke
001.043M 000.999M 0.999M
01.0384M 00.9989M
1.03842M 0.99870M
0.5M
2W 4W Fluke
000.520M 000.499M 0.4989M
00.5188M 00.4990M
0.51878M 0.49894M
Did some tests with higher values:
30M
2W JM101 4W JM101 Fluke
OVLD 13.12V 29.9
595M 10.15V 29.91M
035.391M 2.541V 029.990M 2.541V
10M
2W JM101 4W JM101 Fluke
10.3871M 3.383V 09.9901M 3.383V 9.98M
010.818M 1.693V 009.993M 1.693V
The 4W mode has some changes when I change ranges, the 2W is way off.
What do you think, could the cause be wrong calibration?
Best regards