I just purchased a Fluke 8842 DMM from eBay. It is functional, and seems to produce good readings. I'm working on getting some calibration standards together to give it a good test, but in general it seems fully functional...
But I'm noticing strange behavior, and I wonder if it is a common/known/normal behavior. When powered up with an open corcuit on the test leads, it comes up in DC volts auto ranging mode, and displays a small positive voltage, typically around .4mv DC. The display updates at about .5hz, and the values change in negative increments until it reaches zero, then starts reading negative voltages. If you short the leads, it jumps to near zero (.005mv dc typically) But left open, the display updates every half second, drifting slowly toward more and more negative values, slowing down on a gradual basis. After a couple hours it is around -50mv and after 20 hours it is at .800 mv with the drift slowing down to less than .001 mv change every 30 seconds or so.
It is acting like there is an internal voltage source just above .8v slowly charging a capacitor over the course of 20 hours. I've seen DMM's that drift around when the probes are open, but they usually just display random low values, they don't usually have a directional drift, and .8v is an awfully high and specific value. If I place it in a circuit, the measurements are rock solid and look accurate. So I'm happy enough with the meter, until I get my reference standards in to see how well calibrated it is...
But my question is this open circuit slow drift to -.8vdc, is that normal? Or is that something someone has seen before?
Thanks!