Hi Kleinstein,
I forgot to mention that the supply was on for several hours before and was in use at that specific voltage for other tasks so it was sufficiently warm and I also confirmed with another multimeter that the power supply itself was not drifting upwards
Regarding the current test, I set my supply to 10V to feed a 8.10 kΩ resistor, and let it sit for 2 hours, afterwards the resistor measured 8.11 kΩ, to draw about 1.23 mA of current. I then put the 121GW in series with the resistor in current mode and it measured 1.39 mA at the start 1.43 mA after 1 min, 1.65 mA after 2 minutes, 1.74 mA after 3 min, 1.79 mA after 4 min, and 2.94 mA after pulling out the probes from the multimeter.
I confirmed with another meter that the current was in fact exactly 1.23 mA.
I then checked the signal path through the mA range, namely R43, SHUNT, FB2, R126, R46, R106, FB3 and R104 (refer to the 121GW circuit diagram). All the resistors measured their correct values and the ferrite beads measured shorts.
I then measured the input resistance in the volts mode and it started out at like 2 MΩ and then dropped to like 500 kΩ, I did the same test with my other meter the BM235 and it measured like 4 MΩ and then later dropped to like 1 MΩ so I'm not sure if I did it right or not.
Thanks,
Sojhal