The resistor left of the PTC also looks blown.
They all look in a sad state, they've been bent or squashed by the rear cover or whatever storm this DMM has been through, the middle 1k resistor had a weak solder on the short leg. They all measure good and AFAIK are only related to the V/
input.
Not seeing anything obviously wrong after the diodes, considering I was measuring 60
at the socket for the 400mA range (that's too much power for any of the Melf resistors in there, in fact too much power for all of them added...), I turned towards the suspected previous butchery near the 10A fuse.
A 1/4W 💯
T.H. resistor seemed to have been added over R51. Removing resistors to check them I found that the R51 Melf was open, original value is 1
Replacing that with the right value (I have no 1
precision resistors at hand so I selected value using my Fluke 1520 on low Ohms) is giving me readings quite close to normal. Unfortunately there doesn't seem to be a trimmer for Amps calibration so if I want it closer I'll have to parallel some kOhms...
Resistance measured at the sockets is now 10
, probably right, still I find Extech have been a little optimistic considering that over 2W of power are put into the 9
Melf resistor R53 before the 500mA fuse breaks.
Now I still seem to have a bad contact problem to solve somewhere, let's try cleaning the switch and reflowing the ES51966 IC.