When I set my Extech MM570 to Diode Check mode, which shares a rotary switch position with Capacitance, it shows .407 instead of OL with nothing connected to the com/V inputs. When measuring a known-good rectifier I get .0073
All other functions of the meter check out fine... AC/DC volts/millivolts, Ohms, Capacitance, Frequency, Current - and the readings are bang-on when compared to my Fluke 87
Background Info:
The MM570 a version of the Brymen 859, and is similar to the Brymen 857 (different rms chip/fuses). There's a Q16 NPN transistor just above the rotary switch that's in the path when you select the Diode/Cap range. It seems like it shorts it's collector to ground when turned on, but I'm not sure if it's turned on for Diode or Cap mode. (I don't really want to pull it off until I'm more confident about its role).
Here's a pic of my 570 with the shield removed and some traces for Diode Check mode drawn in:

**I checked out some other Brymen pics from nanofrog, mjlorton, grapsus, and bdunham7... The Extech A, the Amprobe 140-A/160-A, and the Brymen S models do not seem to have Q16... or at least it has a different location/id. I thought it might only be on boards that have Diode/Cap on the same rotary position (and are toggled with the select button)
Other tid-bits:
1. when set to diode mode, the meter shows .407 and draws 5.6mA from the battery with nothing connected to the inputs... the Diode check current on this unit happens to be .4mA, which seems coincidental at least?
2. when powered off and set to diode mode, the resistance from com to V (checked with another meter) is about 4.2M
3. the input impedance measured for DC volts is 10-11M depending on the range (seems right)
4. all the MOVs and PTCs check out, the big blue input cap checks out (i would expect other issues if any of these were bad though)
5. continuity through the range switch assembly seems to be ok
bdunham7 posted a datasheet for the FS970X that might be the basis of Brymen's chip... much of the reference design matches the mm570 but unfortunately the Diode check does not

Anybody got any ideas? or is anybody with a similar meter with Q16 willing to check the current draw and/or resistance from com to V when in Diode mode?
[edited: updated photo showing path through C15]