After some kind soul uploaded the manual to bitsavers, spent a day or so debugging the step generator fault and a few other items. It uses 4 bit binary counter, with scaled resistor D-A values on the outputs to generate the step waveform, driven by clipped ac from the 45 volt regulator transformer winding. Everything looked ok on the counter stages, but not counting. In circuit be and cb junctions looked ok, but deceptive. Replaced both transistors on the first stage flip flop, and it all sprang into life. Plastic transistors used throughout the board, and supply rail is 45v, with 40 v swing on the collectors, so replaced every one with 2n2222a metal can. Also changed the 68k 2w composition resistors to metal oxide, as they run hot to touch. The 225 volt regulator series device, the original fault, ran too hot to touch, around 110 C, so fitted a small fan on card cage to keep things cool. Temp now is around 40 C, which is fine.
It's not a fault of the original design. This unit was switchable 115/230 volts, whereas line voltage here in the uk is usually 245 volts, a nominal 240. This puts added strain on series regulators and probably accounts for the initial failure. Have seen this on some other US kit, the HP5370 counter and the Guildline 9340 thermometer, for example. On the latter, used a heat sink resistor bolted to the chasssis to lose some volts on the input, but it can be a real problem for older US kit run here in the uk. More modern switch mode psu's are no problem.
Just need to finish calibration and should be good for anther 10 or 20 years...