As for the 121GW I had 2 issues. One was that sometimes the meter won't measure capacitance. But only when they are in a (obviously un-powered) circuit. Other DMMs can, though...I guess it depends on how much leakage path (in the form of resistance) there is across the capacitor vs. the current output of the DMM. But let's leave that - the main issue are the AC measurements in mV mode. If there is a DC offset then the readout is wrong. I'm starting to think that maybe the AC coupling is done in firmware or something? Seems to happen when the signal exceeds 2V or thereabouts.
https://youtu.be/G6P6W2s4etA
Well, there are various methods inside DMMs and LCR meters, to measure the capacitance.. and some are able to measure something in-circuit, others maybe even something correct, but most methods require to test the capacitor out-of-circuit.
So your statements, that it does not measure capacitance, or not correctly , or other DMMs can measure inside a circuit, is too generalized, at least it's an inaccurate description of the problem and the environmental condition.
I would not call that a defect, but simply a wrong usage of the DMM.
It's the same story with ACV measurements.. First you claim, that ACV does not work (..at all..), only zeros.
After watching your video, it's obvious, that ACV works correctly... 100mV
pp is measured correctly as about 36mV
rms.
But as as soon as you apply DC of several volts (!!!) to the 50mV AC range, which is not intended at all (AC+DC for 5V and upwards only!), the opamps and/or the digitizing type AC/DC converter in the Hynix chipset is/are completely saturated, displaying zero Volt
AC... 4V
DC is a 50 fold overload on the 50mV
AC range.. good for you, that the 121GW is well protected.
It's clear, that if you reverse this illegitimate DC signal, you'll get a different failure reading.
If I remember correctly, there already had been some discussion on the behavior of measurements on AC+DC signals.. because it has no proper overload indicator for DC, and the allowed DC and AC amplitudes are not specified correctly .. and yes, the manual is not detailed at all at this point.
Maybe you test your 121GW first under proper conditions, before you send it to Dave.