Author Topic: Faulty resistance range Velleman DMM  (Read 8136 times)
Offline Blaffetuur
« on: June 06, 2016, 02:48:34 pm »
I have been using my Velleman DVM9912 for some years now but recently noticed that my resistance measurements are wrong. These are the things I observed:
-All functions except resistance work (checked against my dad's fluke)
-Without anything connected I measure 4k8 ish
-There is a PTC near the righthand connector on the board and when I hold it to warm it up I see the resistance measurement go up a few hunderd ohms
I don't remember what I've done wrong, one thing that could have happend is that I turned off the meter while measuring current.
Do you guys have any suggestions as to where to look ?
(And what could be the function of that PTC ? Is it a protection device or does it some compensation for the temperature measurement ?)
Here are some more pictures: http://imgur.com/a/YVxzI
« Last Edit: June 06, 2016, 03:30:56 pm by Blaffetuur »
Offline retiredcaps
Re: Faulty resistance range Velleman DMM
« Reply #1 on: June 06, 2016, 02:57:03 pm »
Your Velleman looks like it was made by CEM Instruments.
Using the Fluke, measure the resistance of the PTC.  It should be around 1k ohm.  Please post your reading.
The PTC is for input protection.  See
« Last Edit: June 06, 2016, 02:58:53 pm by retiredcaps »
Offline retiredcaps
« Reply #2 on: June 06, 2016, 03:00:32 pm »
If the probes are shorted, do you get 0.0 ohms on the meter?
Offline Blaffetuur
« Reply #3 on: June 06, 2016, 03:04:35 pm »
The PTC measures 890 ohm
And with probes shorted I get 0 ohms
Offline retiredcaps
« Reply #4 on: June 06, 2016, 03:06:28 pm »
PTC reading seems okay.
Can you post clear focused pictures of the entire pcb, both sides please?
Offline Blaffetuur
« Reply #5 on: June 06, 2016, 03:28:22 pm »
Here are some more photos
Couldn't find my sony camera so my phone had to do  :P
Offline retiredcaps
« Reply #6 on: June 06, 2016, 03:35:13 pm »
Pretty good for the phone camera.
Your range switch definitely needs a good clean.  See
http://mrmodemhead.com/blog/fluke-83-iii-rotary-switch/
Offline retiredcaps
« Reply #7 on: June 06, 2016, 03:38:01 pm »
Hmm, did that flying capacitor across the main COB come from the factory?  Same for the yellow bodge wire?
CNR1 is a MOV.  Can you use the Fluke to measure its resistance?  It should measure 0L, open circuit.
edit: pictures saved for future viewers.
« Last Edit: June 06, 2016, 03:58:36 pm by retiredcaps »
Offline retiredcaps
« Reply #8 on: June 06, 2016, 03:44:39 pm »
It looks like version 3 of this board removes the flying capacitor.  Picture in the
https://www.eevblog.com/forum/testgear/help-required-who-is-the-oem-of-this-multimeter/msg265980/#msg265980
Offline Blaffetuur
« Reply #9 on: June 06, 2016, 03:47:31 pm »
This is the first time I open her up so yeah  :-DMM
I measure 1k across the MOV  :o
Offline retiredcaps
« Reply #10 on: June 06, 2016, 03:52:41 pm »
1) Desolder the MOV and measure it out of circuit.
2) Measure the solder points on the pcb after the MOV as been removed.
3) With MOV removed, retest the ohms measurements.
Report all of the above.
Offline retiredcaps
« Reply #11 on: June 06, 2016, 03:56:20 pm »
« Last Edit: June 06, 2016, 03:59:17 pm by retiredcaps »
Offline retiredcaps
« Reply #12 on: June 06, 2016, 04:03:55 pm »
I can't quite read the part number of CNR1?  Can you post it?
Offline Blaffetuur
« Reply #13 on: June 06, 2016, 04:10:38 pm »
Doesn't make a difference, desolderd MOV measures OL, pcb measures 1k and without the MOV the meter still reads 4k8 ...
Offline Blaffetuur
« Reply #14 on: June 06, 2016, 04:11:43 pm »
VDR
07D270K
Offline retiredcaps
« Reply #15 on: June 06, 2016, 04:14:03 pm »
Okay, you can put the MOV back in.  I will have a look at the MOV datasheet.
Have you cleaned the range switch yet?
Offline Blaffetuur
« Reply #16 on: June 06, 2016, 04:17:44 pm »
Yes, cleaned it up with some kontakt spray but didn't solve it
Offline retiredcaps
« Reply #17 on: June 06, 2016, 04:31:32 pm »
I use 91% IPA and scrub with a toothbrush.
Offline retiredcaps
« Reply #18 on: June 06, 2016, 04:36:07 pm »
VDR
07D270K
Hmm, this MOV is only 18V.
http://pdf1.alldatasheet.com/datasheet-pdf/view/553484/BOURNS/MOV-07D270K.html
It must be doing something else than what I thought.  Normally the MOVs are located close to the input jacks, but with the CEM spaghetti layout, I thought it was reasonable to assume this MOV was part of the input protection.
As I mentioned in the Fluke 89IV thread, I need a break.  I'm participating in 3 repair multimeter threads at the same time and it is all starting to blur together.  I'm also gone for a few hours later in the afternoon.
The following users thanked this post: Blaffetuur
Offline Blaffetuur
« Reply #19 on: June 06, 2016, 05:03:03 pm »
VDR
07D270K
I appreciate your help alot ! I'm the asking party so you really don't need to justify your effort  :-+
Offline retiredcaps
« Reply #20 on: June 06, 2016, 07:04:40 pm »
Hmm, it could be nothing, but R7 and R8 in your picture look like they are touching? Circled in red.
If they are touching, separate them slightly.  I don't think this will solve the ohms problem, however.
Offline retiredcaps
« Reply #21 on: June 06, 2016, 07:06:18 pm »
Can you replicate the two readings in this thread
https://www.eevblog.com/forum/repair/fluke-89-iv-powers-up-measures-nothing/msg956417/#msg956417
I only need pictures #2 (1354) and #3 (1355) replicated.  Your Dad's Fluke = Fluke 77 III and your Velleman = 187.
Offline retiredcaps
« Reply #22 on: June 06, 2016, 07:19:57 pm »
I forgot to say please, do the tests above with auto ranging first and then do the same going through the manual range.  I want to see what your meter output/sources through the various ranges.
Offline retiredcaps
« Reply #23 on: June 06, 2016, 07:26:46 pm »
Can we also get a picture of the pcb with the LCD removed?  It looks like you need to remove the 4 screws to release the LCD.  With all the bodges, let's see if there is anything interesting there?
Offline retiredcaps
« Reply #24 on: June 06, 2016, 07:51:45 pm »
Sorry for the multiple posts, but more ideas coming to me now after a little break.
Does your meter show 0L or open circuit with probes removed in following scenarios?
1) manual 400 ohm range
2) manual 4k ohm range
3) manual 40k ohm range
4) manual 400k ohm range
5) manual 4M ohm range
6) manual 40M ohm range
Offline Blaffetuur
« Reply #25 on: June 06, 2016, 08:13:49 pm »
I will check al these things wednesday or thursday, I have a pretty tough exam thursday :-\ so I will see when I have time !
Offline retiredcaps
« Reply #26 on: June 06, 2016, 08:22:37 pm »
Okay. Good luck on your exam.  Maybe after a couple day break, the answer to this problem will become obvious to me or someone else reading this thread.  Posting your findings on Friday is good as well.
Offline Twoflower
« Reply #27 on: June 06, 2016, 08:47:44 pm »
It seems that there is a short between R7 and R8(?). What happens if you separate them?
Argh! too slow :-[
« Last Edit: June 06, 2016, 08:56:11 pm by Twoflower »
Offline retiredcaps
« Reply #28 on: June 06, 2016, 08:58:54 pm »
Argh! too slow :-[
At least your photo shows a better angle than mine.  BTW, if even that doesn't solve the problem, it is a representation of CEM Instruments assembly quality. Or lack of it.
Offline Blaffetuur
« Reply #29 on: June 07, 2016, 01:17:36 pm »
I alo noticed and seperated them but no improvement
Here are my measurements:
RANGE       FLUKE
auto           148k1
               0,145V
               0,22mA
000,0R         38k26
               0,369V
               0,23mA
0,000k         38k26
               0,369V
               0,22mA
00,00k         18k49
               0,145V
               0,05mA
000,0k         7k51
               0,020V
               0,02mA
0,000M         5k9
               0,002V
               0,01mA
00,00M         5k73
               0,2mV
               0,01mA
Here is a high quality picture :
I did remove the LCD and boy is it a mess  :o! I tought some solderpoints were bad, but this takes the cake :wtf:.
Might as well clean it up a little,€5 Velleman to the to the rescue  ::)!
Well it isn't perfect, but a lot cleaner than before, sadly the problem isn't gone
Offline Twoflower
« Reply #30 on: June 07, 2016, 01:34:50 pm »
What a mess! :palm:
The yellow cable looks very strange. They cut the GND plane and replaced it with the wire. It seems they had a offset voltage to compensate. Can you check if the cable at the damaged place is a) still insulated and b) still conductive from one end to the other?
Offline Blaffetuur
« Reply #31 on: June 07, 2016, 01:40:10 pm »
What a mess! :palm:
As bad as that cable looks, its still okay
Offline ModemHead
  • No user-serviceable parts inside.
    • Mr. ModemHead
« Reply #32 on: June 07, 2016, 02:35:24 pm »
Wow.  I've never seen a mess that bad even inside a Hazard Fraught freebie.  :scared:
This is a complete stab in the dark, but if you look around the board you should find some transistors, usually NPN, with the B-C terminals shorted.  These are used as voltage clamps in various places by taking advantage of the reverse avalanche property of the E-B junction.  Check their B-E junctions and make sure they act like diodes.  Keep in mind you may find transistors in normal configuration, or dual-diode arrays also in SOT-23 packages, ignore them.
Don't get caught up with this awesomely-manufactured meter and forget to study for the exam.  :)
Offline Blaffetuur
« Reply #33 on: June 07, 2016, 04:39:13 pm »
I couldn't really concentrate knowing that my poor little meter is waiting for some troubleshooting  :-/O
Offline Blaffetuur
« Reply #34 on: June 07, 2016, 04:41:41 pm »
I will certainly give this a shot  :-+
Offline retiredcaps
« Reply #35 on: June 08, 2016, 02:24:52 am »
When you have time, I would like to get some more data points.
We know the meter doesn't show open circuit, but does show 0.0 when the probes are shorted. If you have a decade resistor box or some resistors, I would like to try one resistor for each range.
For example, measure a 100 ohm resistor in the 400 ohm range, a 1k ohm resistor in the 4k ohm range, etc.  It doesn't have to be 100 ohm, just something between 1 and 399 ohm.