I would carefully check the connections for the channel 2 voltage adjustment. Make sure that everything is well-connected. Next, I would measure the resistance of the variable resistor/potentiometer. I think yours is either open or has a bad spot in it.
While measuring resistance between the wiper to either end, turn it back and forth. You should see a smooth change as you turn it. there should be no sudden jumps or open (infinite resistance) spots. Compare your channel two measurements with channel one. They should be about the same, if the controls are ok.
If you can easily swap parts between channels one and two, you can also try swapping the potentiometers. If the problem moves to channel one, then it's bad. Otherwise, there's something else going on with channel two.
Note: when in tracking mode, channel one's controls sets the voltage for both channels, which is why it works properly.
I hope this helps.
Thanks desy,
I removed the pot, and checked - it is a single turn 100K, in parallel with a fixed 15K resistor. I replaced the pot with a new one just because it wasn't perfectly smooth, but that didn't help. I also traced all the connections on the voltage adjustment side, and there seems to be no issue. here is also a simple test I did, just to see if it's an intermittent connection:
1. I switched to individual mode. Then turned on the supply. the tapped and wiggled wires- the voltage stayed correct and stable.
2. I cranked up the voltage to 60V on channel two, again in individual mode. I waited 3 minutes without touching anything. The voltage started to get unstable, moving between 58 and 63, then after about 5 minutes it jumped up to 98 volts.
3. I switched to tracking. set both voltages to mid-way. Then I switched to individual - the voltage on channel two INSTANTLY (over 2 seconds) shot up to 98.
Each output channel is controlled by a TL494. What could make it the voltage on ch2 rise like that? or become unstable?