Hi all!
I’m chemist working with mass spectrometers (MS), instruments that weight ions really accurately. We have a ~30 year old MS where the electronics are not working quite right... I've done some troubleshooting, but I need help with the last details.
Mass spectrometers consist of an ion source and a detector. The ion source generates an ion beam at low pressure. The beam is accelerated and steered under high vacuum by static high voltage potentials into the detector. Unfortunately, I cannot see any ions in the detector and I'm sure the detector works because it has an internal ion source where I can detect ions. There are many different high voltages that control the path of the ions and among those are two pairs of voltages called XDFL and YDFL that deflect the ions in x- and y-direction to keep the beam in the right path. The software readout of the voltages are ~0 V no matter what I set the voltages to (-200 V to +200V). I have confirmed that the software readout matches measurements with multimeter. The XDFL voltage is also supposed to be pulsed about every second with an additional 18 V so the ion beam is deflected away from the detector when it is weighing the ions.
All of the high voltages are coming from one big PCB with 11 smaller identical PCBs mounted on it, connected with ribbon cables, see the first photo. My impression is that each of the smaller boards generate 3-5 V AC for the electronics underneath that control each of their own DC high voltages. I think it is done this way because some of the voltages float on the 3 kV that accelerate the ions. Coincidentally, the XDFL and YDFL voltages both come from underneath the same smaller PCB so it makes sense neither of them work correctly if something with the shared electronics is wrong. On the other hand, the mounted PCB appears to work correctly and, from what I can tell, the electronics underneath are separated so maybe there is the same fault on both parts of the circuit…

My understanding of the circuitry below, see second photo, is that a microprocessor controlled by software sends a digital signal of the desired voltage. This goes through an optocoupler to a DAC and the analog output is buffered by the OPAMP. There are also SSRs, voltage regulators and diodes, but it is not clear to me how they are connected and generate the +-200 V DC, but I assume some of it is generating 5 V DC for the logic from the 5 V AC from the transformers above.

When it comes to comparing to the parts that work (e.g. DPL4 voltages), when the voltage is set in software too 100 V, the OPAMP gets V+=8.8 V, V-=0 V, IN+=1.4 V, IN-=1.4 V and OUT=4.0 V, but for the faulty circuits V+=8.8 V, V-=0 V, IN+=1.4 V, IN-=8.2 V and OUT=0 V. IN- is connected to the AQW210EH SSR and K2700 voltage regulator. Does this indicate that the SSR or voltage regulator is toast? How can I test them?
I need to fix the circuitry and would like to know what can have gone wrong and how I can test what is wrong. I have access to oscilloscope and other equipment and can supply more photos or measurements if needed. Any more knowledgeable people that can help out a chemist?
The ICs are from left to right:
PS2501-4:
https://www.renesas.com/us/en/document/dst/ps2501-1-ps2501-4-ps2501l-1-ps2501l-4-data-sheet LTC1456:
https://www.analog.com/media/en/technical-documentation/data-sheets/1456fs.pdf OPA251:
https://www.farnell.com/datasheets/81252.pdf AQW210EH:
https://www3.panasonic.biz/ac/e_download/control/relay/photomos/catalog/semi_eng_ge2a_aqw21_e.pdf K2700:
https://datasheetspdf.com/pdf-file/549187/ToshibaSemiconductor/K2700/178L05:
https://www.ti.com/lit/gpn/UA78L L78L09:
https://www.st.com/resource/en/datasheet/l78l.pdf