Author Topic: Fluke 115 Repair Help  (Read 936 times)

0 Members and 1 Guest are viewing this topic.

Offline hassan58Topic starter

  • Newbie
  • Posts: 2
  • Country: za
Fluke 115 Repair Help
« on: November 29, 2021, 11:48:44 am »
I am looking for the following resistor values for a Fluke 115 that has been damaged
The board has FLUKE-117-3001 REV 008 on it
 Image attached

Need the values for R38, R39, R42, R43, R46

It would be helpful to confirm the values for R37, R40, R41 R44 and R45 as well please
 

Offline rvalente

  • Frequent Contributor
  • **
  • Posts: 726
  • Country: br
Re: Fluke 115 Repair Help
« Reply #1 on: October 12, 2022, 07:26:13 pm »
Hope this helps

Board FLUKE 117-3001 REV 009
Model FLUKE 115
 
The following users thanked this post: mcinque

Offline Ugur

  • Regular Contributor
  • *
  • Posts: 84
  • Country: tr
Re: Fluke 115 Repair Help
« Reply #2 on: October 13, 2022, 10:34:00 am »
R38 : 150K
R39 : 75K
R42 : 75K
R43 : 75K
R46 : 49,9K
 

Offline hassan58Topic starter

  • Newbie
  • Posts: 2
  • Country: za
Re: Fluke 115 Repair Help
« Reply #3 on: October 17, 2022, 05:25:44 am »
Thanks @rvalente and @Ugur
Your help has resolved my problem
 
The following users thanked this post: rvalente, Ugur


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf