Fluke 1507 Problem
« on: August 26, 2022, 04:27:14 pm »
My fluke 1507 isolation tester gives a wrong measurement on Voltage. It must be 0.0 V but on the screen is  -0.1 . Must i calibrate the meter or is there something broken.
 

Offline Black Phoenix

  • Super Contributor
  • ***
  • Posts: 1135
  • Country: hk
Re: Fluke 1507 Problem
« Reply #1 on: August 27, 2022, 11:10:17 am »
No need to calibrate. That's normal, residual voltage.

If you insert both probes and touch between them the value should go to zero.
 

Offline Stijn_vTopic starter

  • Newbie
  • Posts: 4
  • Country: be
Re: Fluke 1507 Problem
« Reply #2 on: August 27, 2022, 01:55:31 pm »
No, when you touch the leads together it staysin -0.1 voltage.
 

Offline Stijn_vTopic starter

  • Newbie
  • Posts: 4
  • Country: be
Re: Fluke 1507 Problem
« Reply #3 on: August 27, 2022, 02:06:33 pm »
No, when you touch the leads together it staysin -0.1 voltage.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 1507 Problem
« Reply #4 on: August 27, 2022, 07:22:34 pm »
I used to buy used and abused Fluke meters and one of them, when probes shorted, showed -0.2 ohms  My post from 8.5 years ago.

See post #2 at

"https://www.eevblog.com/forum/repair/multimeter-repair-time-again!-fluke-177-reads-8-ohms/"
« Last Edit: August 28, 2022, 12:03:15 am by retiredcaps »
 

Offline J-R

  • Super Contributor
  • ***
  • Posts: 1283
  • Country: us
Re: Fluke 1507 Problem
« Reply #5 on: August 27, 2022, 10:28:50 pm »
Voltage measurement accuracy specifications for that device are: 50 Hz to 400 Hz ± (% of Rdg + Digits): ±(2% + 3).

So best case the device shouldn't be trusted to within 0.3V anyway.

Supply some known inputs and see if they are correct?
 

Offline Stijn_vTopic starter

  • Newbie
  • Posts: 4
  • Country: be
Re: Fluke 1507 Problem
« Reply #6 on: August 29, 2022, 09:35:55 am »
Must i clean the PCB? There is nothing dirt on the pcb ?
 

Offline CapLeaker

  • Contributor
  • Posts: 45
  • Country: ca
Re: Fluke 1507 Problem
« Reply #7 on: August 29, 2022, 10:42:59 am »
Cleaning the board with 99% alcohol is a cheap investment. I’ d do the same thing if it would be mine. retiredcaps fixed his by cleaning the board twice. Hopefully it will fix yours too…
I’ve got a Fluke 178 when they came out over 20 years ago. So far so good, never missed a beat and is still within spec.
 

Offline J-R

  • Super Contributor
  • ***
  • Posts: 1283
  • Country: us
Re: Fluke 1507 Problem
« Reply #8 on: August 29, 2022, 11:15:50 pm »
There is nothing wrong with it, why "fix" it until it's really broken??
 