Are you absolutely sure that the replacement fuses are good? Have you tested them with another meter? If you don't have another meter, can you take them out and test them each with the Fluke 179 in ohms mode?
When you put the meter in the ohms mode, and connect a probe directly from the red V+ socket to the 400mA current socket, what does the display read? How about when you connect the probe directly from the red V+ socket to the 10A current socket? (If the meter and fuses are good, readings in the neighborhood of 1.8 ohm for the 400mA socket and 0.1 ohm for the 10A socket are typical.)