Hi to all,
I need some help for reparing half dead Fluke 187.
I normally service electronic devices, but trying to repair this dmm gives me a headache. I have a strange problem with this instrument. When it is turned on, no matter which measuring range is set it always just writes OL and the bargraph is always at maximum and does not react to measurements as if the test probes are not connected.
Another problem is that when switching to continuity mode it always beeps constantly without or with plugged test leads. Here are the results of some of my measurements on it:
1. All measuring ranges can be selected and all this is neatly displayed on the screen.
2. Everything is reset to factory settings but no change.
3. All terminals are clean and in the measuring range for uA and mA it correctly detects whether the test leads are plugged or not.
4. All protective resistors and other elements on the input side are measured and ok.
These are the measured resistance values with the other instrument when the probes are connected to its COM + Resistor, Voltage, Diode terminals while it is off.
5. Vac-0 Ohm
6. mVac-0 Ohm
7. Vdc-0 Ohm
8. mVdc-0 Ohm
9. Ohm, nS, Con.-1 Mega Ohm.
10. Diode, Cap.-1.47 Mega Ohm.
11. Temp.-1.48 Mega Ohm.
12. mAac-0 Ohm.
13. uAac-0 Ohm.
14. mA-0 Ohm.
15. mAdc-0 Ohm.
16. uAdc-0 Ohm.
All contacts on the rotary switch are clean and ok.
The thin white resistive network is tested and all values are ok.
There are no traces of dirt or corrosion but I just don't know what could be the cause ...? I ask for any help or advice.
P.S. sorry in my english.