No, commands exists in both versions.
I believe the difference between V1.xx and V2.xx lies in the hardware.
The LDF for V1.xx does not include an initialization command to enable the MMU.
V2.xx does include it.
This is likely because the increase in firmware size necessitated hardware changes, and with the installation of a larger-capacity flash memory, enabling the MMU became unavoidable.
The firmware that does not enable the MMU is found in the first-generation 19x equipped with UHM V2.1.
Also, the memory test section of the 123 service manual describes the response to the ID command in UHM mode. The manual shows V3.0, so his 123 must be an older model.
It might be difficult unless we know exactly what commands the V1.xx firmware updater sends. If we could use the RB/GB commands available in UHM V3.0, we could at least check the calibration status.
123 V108 and V203