All TDS600 and TDS700 series scopes are essentially capable of 1GHz, the limiting factors are personality identification resistors on the acquisition board that tell the processor board what it is, and a capacitor on each input diff pair on the 500MHz model, located between the hybrids and acquisition chips that act as a roughly 500MHz low-pass filter.
All hybrids are capable of 1GHz, and essentially all 500MHz acquisition boards that have a 1GHz equivalent can be modded to 1GHz (This may not be true for the early non A, B, C, D models, I haven't played with them enough to check).
I.e., any TDS754A, B, C, D can be upgraded to a fully functional TDS784 of the same letter type, 'simply' by swapping the ID resistors, removing the bandwidth limiting capacitors, and then running a full cal/adjustment procedure. (500MHz scopes are not calibrated from factory to 1GHz. The scope will work at 1GHz, but it won't be calibrated beyond 500MHz).